Re: UT transducer connecting

Posted by: Linas Svilainis Profile (PID_197), E-mail: Address, on June 16, 2008 at 09:52 :

In Reply to: UT transducer connecting posted by : Chad , E-mail: Address, on June 12, 2008 at 22:39 :

----------- Start Original Message -----------
: I seem to recall being told one time that you should not connect your transducer to your flaw detector while the machine is on. Is this correct or does it not matter one way or the other?
------------ End Original Message ------------
If you are speaking about passive probe, this requirement is probably for safety issues.
1.The connector (BNC,SMA ) that is usually used to connect transducer to equipment is not designed for "pre-mating earth".
2.Not all equipment have a galvanic isolation of "hot" center tap.
This means that there is a possibility to connect a center tap first. In such case the operator can get electrical shock.
Speaking of active probe (with preamp inside or array) this requirement is applicable for equipment that is not designed for "hot plug".
Despite everything said people DO CONNECT transducers while equipment is on and no one was seriously hurt or equipment damaged.
My advise would be to follow this rule.

regards,

Linas