Re: UT for below 8mm thk

Posted by: Tom Nelligan Profile   Exhibitors stand (PID_139), E-mail: Address, on October 27, 2008 at 16:54 :

In Reply to: Re: UT for below 8mm thk posted by : David Harvey Profile , E-mail: Address, on October 27, 2008 at 16:12 :

Good near surface resolution is a function of transducer selection as well as test material properties. For straight beam testing or thickness gaging in steel, a 20 MHz broadband transducer -- within the bandwidth of most portable flaw detectors -- will typically resolve reflectors within 0.5 mm of the contact surface, and a 10 MHz broadband transducer will typically get within 1 mm or better. But if you don't have a high frequency transducer available, I agree that there will be limitations. It's just a question of using an optimum transducer and instrument setup.

----------- Start Original Message -----------
:: If near-surface resolution is a concern, material this thin can be a tough one. You can find yourself in a spot where you will need to resort to a through transmission type of test, which won't much care about surface resolution.
------------ End Original Message ------------