Re: calibration

Posted by: Tom Nelligan Profile   Exhibitors stand (PID_139), E-mail: Address, on November 18, 2008 at 17:18 :

In Reply to: calibration posted by : LANRE ALAPINI Profile , E-mail: Address, on November 18, 2008 at 16:49 :

Yes, there will normally be a very significant difference in the zero offset calibration between a single element contact transducer and a dual element transducer of similar frequency and diameter. In the case of a dual, you have to zero out the pulse transit time through the internal delay line, while with a contact transducer you are zeroing out only a very thin wearplate. (Other minor factors like cable length and instrument switching delays wouldn't normally vary with transducer type.)

Material sound velocity is of course independent of tranducer type.


----------- Start Original Message -----------
: this kind of question could be could there be any diferent between the calibration of a single element transducer and dual element transducer
------------ End Original Message ------------