Re: Urgent- ToFD for Steam Reformer Tube
Posted by: Michel Couture(PID_2805), E-mail: Address, on November 30, 2008 at 20:10 :
In Reply to: Urgent- ToFD for Steam Reformer Tube posted by : farabirazy , E-mail: Address, on November 29, 2008 at 06:05 :
Abi,
The frequency you are using is too high. ASTM A388 para 8.2.1 states that the prefered frequency for coarse grain austenitic materials is 1 MHz and that in many instances it may be necessary to use a frequency of 0.4 MHz. I know that ASTM A388 is for Ultrasonic Examination of Heavy Steel Forgings, but the reason for freqyency selection doesn't change. You are dealing with a very coarse grain structure and you need to lower your frequency..
Hope this will help.
Cheerio's
Michel----------- Start Original Message -----------
: Dear Gentlemen,
: Does anyone ever involve in research of ToFD for Cast austenitic stainless steel used as ammonia reformer tube? I tried to detect artificial defect in IN 519 material, but I can not distinguish the backwall signal. Can anyone explain why and what should I do to detect the defect in this material by using ToFD? As FYI, I used 2.25 MHz ToFD transducer. Is there any commercial ToFD transducer with frequency below 2.25 MHz?
: Please help me...
: Sincerely Yours,
: Abi
------------ End Original Message ------------