Re: penetrameter requirement

Posted by: DJ Kallhof Profile (PID_1317), E-mail: Address, on January 08, 2009 at 17:57 :

In Reply to: penetrameter requirement posted by : susan , E-mail: Address, on January 08, 2009 at 00:50 :

At least one engine manufacturer allows steel to be used in place of cobalt but I am not sure about nickel. A softer or less dense material may be used. The SPM I checked went so far to say use the penetrameter designated for the next greater part thickness if you are using a iron based penetrameter for nickel or cobalt. I checked the density of cobalt and nickel and they are nearly identical. Unfortunately, an auditor who doesn't understand the process could argue that nickel is actually a little denser and would disqualify the technique.

----------- Start Original Message -----------
: If I have a coblat material to x-ray but do not have any cobalt penes, is it acceptable to use Nickel penes? This is in regard to the aerospace industry.
------------ End Original Message ------------