09:12 Mar-06-2013
Ultrasonic testing in high temperautre

Hi
I am new inspector UT/2, I will want to know how to make the
calibration for measurement from thickness on pipe at high
temperature, my apparatus is olympus EPOCH LTC with daul probe (400°C)
and olympus GM2 plase help me
best regards

 
 Reply 
 
mahmoud refaat
NDT Inspector,
Egypt, Joined Aug 2012, 16

mahmoud refaat

NDT Inspector,
Egypt,
Joined Aug 2012
16
11:02 Mar-06-2013
Re: Ultrasonic testing in high temperautre
In Reply to hamzakhan at 09:12 Mar-06-2013 (Opening).

HI, hamzakhan
at first i want to clarify that i am junior like you but have little experience
1-first based on ASME V the calibration block will be pipe or can use flat block depending on the pipe diameter so what is the pipe diameter and what is the code ?
2- the calibration steps in general will be using the calibration block if it is flat or curvature one depending on pipe diameter and you must check about the examined material P number and must be like the Calibrated bloc P number
3- the temperature according to ASME V the allowable difference between calibration block and examined temperature with in 14 degree Celsius but i think you must check about the data sheet for the used probe at first and if it is fit to be used in that temperature or not and if it is O.K so you can use it with in the temperature limit mentioned in the Probe data sheet

 
 Reply 
 
John
John
14:17 Mar-06-2013
Re: Ultrasonic testing in high temperautre
In Reply to mahmoud refaat at 11:02 Mar-06-2013 .

As you mentioned, this is for thks, just make sure the step wedge is the same temp as the part inspected
Do this by placing step wedge on part to be inspected

 
 Reply 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
14:57 Mar-06-2013
Re: Ultrasonic testing in high temperautre
In Reply to hamzakhan at 09:12 Mar-06-2013 (Opening).

As noted above, you need to calibrate velocity and zero on a test block whose temperature matches that of your test piece. Also, if you are using one of our high temperature duals at 400°C, you need to keep contact with the hot surface brief. Please see our application note on this subject for more details:

http://www.olympus-ims.com/en/applications/high-temperature-ultrasonic-testing/347-pos.20.html

 
 Reply 
 
mahmoud refaat
NDT Inspector,
Egypt, Joined Aug 2012, 16

mahmoud refaat

NDT Inspector,
Egypt,
Joined Aug 2012
16
22:55 Mar-06-2013
Re: Ultrasonic testing in high temperautre
In Reply to Tom Nelligan at 14:57 Mar-06-2013 .

sorry i thought that you will use flaw detector not thickness measurement
as noted above the procedure will be
1- check for probe data sheet
2- i think at first check that the material is the same for the piping material or at leat the same P no
3- elevate the temperature of the step wedge to the examined temperature
4- so now the material and temperature like the examined material so the attenuation and the scattering inside the material will be the same as the examined material (in addition to dv/dT temperature effect we already took it in our consideration in calibration step)
5- use calibration using two calibrating steps thin and thick one and you must to check that the examined thickness located between the two steps i.e if the piping thickness 0.75 inch so use calibration on 0.25 and 0.1 inched or use one step with the same thickness i.e 0.75 inch

Finally ASME V doesn't talk as for temperature effect so if any one found the article or clause for the correction factor or any note regarding high temperature examination advise me

 
 Reply 
 
hamzakhan
NDT Inspector,
E N A CT, Algeria, Joined Mar 2013, 4

hamzakhan

NDT Inspector,
E N A CT,
Algeria,
Joined Mar 2013
4
14:24 Mar-09-2013
Re: Ultrasonic testing in high temperautre
In Reply to hamzakhan at 09:12 Mar-06-2013 (Opening).

thank yuo for all,
the temperature variation in speed-on U.S. it has a significant effect or no

1-my job is to measure the thickness of a pipe 6 "duplex steel material SS A790
nominal THK: 11.5 mm and MIN THK: 10 mm (12.5%)
2-my block calibration is V1 ( carbon steel)
3-operating temperature of the pipe is 125°C-225°C and  second is 400°C

tell me how to do the calibration and measurement in this case i'm lost

 
 Reply 
 
Sudheer Jai Krishnan
Sudheer Jai Krishnan
18:58 Mar-09-2013
Re: Ultrasonic testing in high temperautre
Hello Hamzakkan

You are having a SS material and you calibration block is a carbon steel which is not correct. You have to get a similar material as a cal block. Also, if you have a 38DL Plus with you, then you will not face any difficulties when you do the High Temperature thickness measurements as 38DL Plus has got the facility of entering the material temperature in it. Epoch LTC what you are using is a dedicated hand held Ultrasonic Flaw Detector and not a thickness gage.

Regards

Sudheer Jai Krishnan
 
 Reply 
 
hamzakhan
NDT Inspector,
E N A CT, Algeria, Joined Mar 2013, 4

hamzakhan

NDT Inspector,
E N A CT,
Algeria,
Joined Mar 2013
4
09:15 Mar-10-2013
Re: Ultrasonic testing in high temperautre
In Reply to Sudheer Jai Krishnan at 18:58 Mar-09-2013 .

hello and thank you very much
In my situation with EPOCH LT / EPOCH LTC what is the solution
I made a wedge duplex SS in the workshop 5mm-10mm-20mm ,but the problem is I do not know how I correct speed to high temperature there a formula V = f (T) and
later I recommend 38DL for my company

Best regards

 
 Reply 
 
John
John
15:43 Mar-10-2013
Re: Ultrasonic testing in high temperautre
In Reply to hamzakhan at 09:15 Mar-10-2013 .

The formula is very simple, but I am doubting you would remember as you don't need one if the block is the same temp as the material.(as stated on this thread several times). Most codes also state "shall" when it comes to the test piece & cal block at same temp, which makes formulae a moot point.

 
 Reply 
 
mahmoud refaat
NDT Inspector,
Egypt, Joined Aug 2012, 16

mahmoud refaat

NDT Inspector,
Egypt,
Joined Aug 2012
16
00:57 Mar-11-2013
Re: Ultrasonic testing in high temperautre
In Reply to John at 15:43 Mar-10-2013 .

hamzakhan
referring to the materials Standard (ASTM Stand and ASME II Part A and ASME IX ) you inspected material SA-790 classified as P n- 10 H group 1 and Carbon is P no1 so you have to use calibration block from the same material or at least from the same P-no
secondly i think regarding the instrument i also recommended 38DL Plus because it's high accuracy regarding thickness measurements in high temperature and you must select suitable transducer for example Gage Dual Transducers and check about the temperature limit for it
finally regarding the variation of the temperature V.s sound speed i am with Mr.john why you need it if you will calibrated the block at the same temperature and you can calculate it reversely from your calibration and if the formula for the effect of the temperature on velocity you can read the next article for the formula :
http://www.ndt.net/article/wt1097/hammond/hammond.htm

 
 Reply 
 
ajith m r
NDT Inspector,
geecy Industrial Services Pvt Ltd, India, Joined Mar 2013, 2

ajith m r

NDT Inspector,
geecy Industrial Services Pvt Ltd,
India,
Joined Mar 2013
2
08:39 Mar-11-2013
Re: Ultrasonic testing in high temperautre
In Reply to hamzakhan at 09:12 Mar-06-2013 (Opening).

Hello,
Am not that exprt in this, but I ask this qus to ma sir, am noting down his suggestion

1) Calibration block must b on same metrial, and temp, temp must b check with infrared thermometer,
And if u use temp stick the error is upto 50 deg, that may distroy ur prob,

2) Coplnt, should withstand up to that temp, water oil and greez wil get evaprt so wil get some gel. It should apply on prob , and the mechine u noted above I think it has an ability to retrive the image, so just place d probe take take it with in some seconds, after a scan give tym to cool dwn ur cristal

3)Bec of high tem make sure ur body , meechine, probe cable r safe.

 
 Reply 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
14:17 Mar-11-2013
Re: Ultrasonic testing in high temperautre
In Reply to hamzakhan at 09:15 Mar-10-2013 .

"the problem is I do not know how I correct speed to high temperature there a formula V = f (T)"

The application note link that I gave you above contains the general answer. Velocity in carbon steel drops by approximately 1% per 55 degrees Celsius. There is a small variation between different alloys, but that number is usually good enough to work with.

1
 
 Reply 
 

Product Spotlight

FAAST-PA! OEM Patented phased Array for high speed UT inspection

Multiangle, Multifocus, Multifrequency, Multibeam. Instead of stacking UT electronics and having m
...
any PA probes, FAAST-PA is able to transmit all delay laws within ONE single shot in Real time.
>

EKOSCAN Phased Array

In order to always fit your needs, EKOSCAN can manufacture any type of UT transducer, either convent
...
ional or Phased Array. As an ISO 9001: 2015 certified company, EKOSCAN is extremely careful as far a material selection and manufacturing processes are concerned. Our probes guarantee our customers the benefits of latest innovations regarding piezo-composite, backing, impedance adaptation layer, etc. Specific probes for hostile environment: high temperature, high pressure, corrosive environment,etc. Specific probes designed to fit your specific application: optimization of every parameter to guarantee you the best detection.
>

A1525 SOLO

A1525 Solo – the most compact and affordable TMF unit with two phased array transducers and 3D v
...
isualization and analysis software in standard delivery set. A compact, ergonomic and easy to handheld Phased Array unit based on Total Focusing Method for easy-going imaging of inspection objects with two-dimensional and three-dimensional visualization and evaluation of inspection results.
>

Cygnus 6+ PRO Multi-Mode Ultrasonic Thickness Gauge

The Cygnus 6+ PRO thickness gauge is the most advance gauge within the Cygnus range with key featu
...
res including: comprehensive data logging; A-scan and B-scan display; manual gain control; Bluetooth connectivity; and much more. With its unique dual display and three measuring modes (Multiple-Echo, Echo-Echo and Single-Echo), this surface thickness gauge offers maximum versatility for inspections.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window