Hello everybody
Today I have a question : When I check screen height linearity and amplitude linearity.If it's not meet the requirement as in ASME V article 4.How can I do next?

 
 Reply 
 
Tariq KHELIFA
NDT Inspector, QC-inspector
South OceaN Inspection Service-SOIS-, Algeria, Joined Dec 2012, 3

Tariq KHELIFA

NDT Inspector, QC-inspector
South OceaN Inspection Service-SOIS-,
Algeria,
Joined Dec 2012
3
12:38 Jul-01-2013
Re: UT-Screen Height Linearity And Control Linearity
In Reply to Lanh Huyet at 05:56 Jul-01-2013 (Opening).

see EN 12668-3 if the process not meet this procedure, must be calibrated again by 12668-1
Regards

 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1286

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1286
14:34 Jul-01-2013
Re: UT-Screen Height Linearity And Control Linearity
In Reply to Lanh Huyet at 05:56 Jul-01-2013 (Opening).

The screen height linearity is rarely an issue in modern instruments with digital display monitors and the 5% screen-height tolerance should be very easy to achieve even with the difficulties of maintaining coupling on the targets suggested by ASME. The Amplitude Control linearity assesses the amplifier performance. In either case, if you cannot achieve the linearity required in ASME Sec. V Art. 4 the operator has no ability to make the necessary adjustments and the unit should be sent to the manufacturer or suitable repair shop to bring the parameters back into the acceptable range of tolerances. Of course, this means that the unit cannot be tagged with an acceptable calibration sticker and must not be used until the corrections are made and acceptable calibration achieved.

 
 Reply 
 
sixxerxxre
sixxerxxre
16:44 Apr-23-2014
Re: UT-Screen Height Linearity And Control Linearity
In Reply to Lanh Huyet at 05:56 Jul-01-2013 (Opening).

hello, in fact i have no answer but a few questions. in nfen 12668-3 the frequency for verification of " screen height linearity and amplitude linearity " is very clear in Table 1 § 3.1 : what is to check daily and every week is also clear and cannot be interpreted.
for the ASME V Code, § T-532 and T-533 mention " shall be performed at the beginning
of each period of extended use (or every 3 months, whichever is less)."
have you any idea of what is an " extended use " ?
does " whichever is less " mean that it needs to be checked frequently ?
what is checked in this verification ? the Flaw Detector with the help of an anonymous probe or a " complete ultrasonic examination system. " ?

 
 Reply 
 
Mario Talarico
NDT Inspector,
Italy, Joined May 2010, 423

Mario Talarico

NDT Inspector,
Italy,
Joined May 2010
423
17:54 Apr-24-2014
Re: UT-Screen Height Linearity And Control Linearity
In Reply to sixxerxxre at 16:44 Apr-23-2014 .

for the logic of what is written "each period of extended use (or every 3 months, whichever is less)" implies that an "extended use" could exceed 3 months (otherwise why ask the question, "whichever is less"! How far can exceed ? Indefinite. But for the same logic can be less than 3 months ("whichever is less"!). How far can exceed ? Indefinite.
So, if you are under fierce verification, I think it is wise to secure for the worst case, attaching to the certificate of verification of the examination Ut "screen height linearity and amplitude linearity" for each probe included in the report, including stright beam.
If I remember correctly, other members in the past have suggested the use of the probes so as to involve the verification of the system, ie probe + instrument (what is called "equipment" in 12668-3.
Greetings
mario

 
 Reply 
 

Product Spotlight

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

Echomac® Small

Available with up to eight channels of electronics to detect and evaluate thickness, flaws and eccen
...
tricity, this UT tester is housed in a convenient, smaller cabinet. This instrument can be used in conjunction with bubbler or immersion tank systems, or with a test bench or in laboratory applications. The Echomac® Small is available in the FD4, FD6 or FD6A versions.
>

EKOSCAN Phased Array

In order to always fit your needs, EKOSCAN can manufacture any type of UT transducer, either convent
...
ional or Phased Array. As an ISO 9001: 2015 certified company, EKOSCAN is extremely careful as far a material selection and manufacturing processes are concerned. Our probes guarantee our customers the benefits of latest innovations regarding piezo-composite, backing, impedance adaptation layer, etc. Specific probes for hostile environment: high temperature, high pressure, corrosive environment,etc. Specific probes designed to fit your specific application: optimization of every parameter to guarantee you the best detection.
>

Cygnus 6+ PRO Multi-Mode Ultrasonic Thickness Gauge

The Cygnus 6+ PRO thickness gauge is the most advance gauge within the Cygnus range with key featu
...
res including: comprehensive data logging; A-scan and B-scan display; manual gain control; Bluetooth connectivity; and much more. With its unique dual display and three measuring modes (Multiple-Echo, Echo-Echo and Single-Echo), this surface thickness gauge offers maximum versatility for inspections.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window