10:17 Jun-20-2016
Dead Zone calculation

There is always a minimum thickness for any particular UT setup where the results are not sensitive,which is behalf of the interference effect close to the transducer,which we call as dead zone
How to calculate the dead zone

If somewhere i am wrong in my concept please rectify me

Thanks to you all in advance

 
 Reply 
 
Anmol Birring
Consultant,
Birring NDE Center, Inc., USA, Joined Aug 2011, 832

Anmol Birring

Consultant,
Birring NDE Center, Inc.,
USA,
Joined Aug 2011
832
11:42 Jun-20-2016
Re: Dead Zone calculation
In Reply to Dhiliban at 10:17 Jun-20-2016 (Opening).

There is no calculation. You just have to measure it.

 
 Reply 
 
massimo carminati
Consultant, AUT specialist
IMG Ultrasuoni Srl, Italy, Joined Apr 2007, 691

massimo carminati

Consultant, AUT specialist
IMG Ultrasuoni Srl,
Italy,
Joined Apr 2007
691
23:08 Jun-20-2016
Re: Dead Zone calculation
In Reply to Dhiliban at 10:17 Jun-20-2016 (Opening).

The dead zone of a UT measuring chain and test condition has to be measured and is the minimum distance from beam entry point where an echo can be distinguished.

1
 
 Reply 
 
Satyajeet Jena
Satyajeet Jena
08:54 Aug-03-2018
Re: Dead Zone calculation
In Reply to massimo carminati at 23:08 Jun-20-2016 .

It is a short time at which the transmitted signal blinds the received signal due to high noise of the crystal. Just like eye blinking. Happens in probe which transmit & receive in one crystal..

Thanks

 
 Reply 
 
langtuteng
Armenia, Joined Nov 2014, 199

langtuteng

Armenia,
Joined Nov 2014
199
03:45 Aug-05-2018
Re: Dead Zone calculation
In Reply to Dhiliban at 10:17 Jun-20-2016 (Opening).

Yes, the dead zone is measured. it is based on your probe, equipment and sensitivity. and it varies from different sensitivity. so I suggest you to measure it under the probe and equipment combined system, also with the sensitivity. the dead zone is the distance where you can distinguish the first echo behined the initial pulse echo, and mostly we define if the trough between the initial pulse wave and first echo wave is 6dB lower than the peak of first echo wave , then we can distiguish them and this diatance is dead zone.

2
 
 Reply 
 
Arman
Engineering,
Turkey, Joined Oct 2014, 42

Arman

Engineering,
Turkey,
Joined Oct 2014
42
08:48 Aug-05-2018
Re: Dead Zone calculation
In Reply to massimo carminati at 23:08 Jun-20-2016 .

Gents,

Thanks for all explanations, could you please explain effect of parameters like probe frequency, crystal diameter ...etc on length of deadzone?

 
 Reply 
 
Paul Holloway
Consultant,
Holloway NDT & Engineering Inc , Canada, Joined Apr 2010, 253

Paul Holloway

Consultant,
Holloway NDT & Engineering Inc ,
Canada,
Joined Apr 2010
253
10:39 Aug-05-2018
Re: Dead Zone calculation
In Reply to Arman at 08:48 Aug-05-2018 .

There was a lengthy discussion regarding dead zone here:

https://www.ndt.net/forum/thread.php?msgID=62184

A simple forum orGoogle search would have revealed this.

1
 
 Reply 
 

Product Spotlight

Magnetic X-Ray Pipeline Crawler 6''-60''

Suitable to work in 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometer
...
s and provide the efficient inspection of the pipeline.
>

Research and Application Development For NDT

Acuren’s Research and Application Development specializes in the development of advanced ultraso
...
nic inspection techniques and systems for challenging inspection applications, with an emphasis on practical solutions which are field deployable. Services include manual and automated ultrasonic inspection system development, inspection technique optimization using laboratory scale studies and ultrasonic modeling (CIVA, BeamTool), preparing technical justification for technique evaluation and qualification (Probability of Detection and sizing accuracy studies), inspection/calibration/analysis procedure preparation to support field deployment of custom techniques, and development of custom imaging algorithms to support challenging inspection applications.
>

NDTkit RT

NDTkit RT, TESTIA's Digital Radiography software The NDTkit product line software for X-ray analysi
...
s. NDTkit RT is a software benefiting from the Ultis kernel which is dedicated to radiographic image analysis. It offers a set of tools and filtering processes to assist RT operators in finding relevant flaws.
>

AIS229 - Multipurpose Real Time System

Latest standard & automatic real time system developed by Balteau. The AIS229 has been designed to
...
do series inspection in a wide variety of industry. Composed of a shielded cabinet, 5 axis manipulator, x-ray generator and tubehead from 160kV to 225kV, a fl at panel & much more, the AIS229 is most certainly one of the most multipurpose RTR system available on the market.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window