16:17 Jul-05-2016
Edge scatter in radiography

The iso 17636 states in its para. 7.5.1
"With Ir 192 and Co 60 radiation sources or in the case of edge scatter, a sheet of lead can be used as a lower energy scattered radiation filter between the object and the cassette. The thickness of this sheet shall be between 0,5 mm and 2 mm in accordance with the penetrated thickness."
Can someone explain what is this scatter and if the standard means really to put 2 mm lead sheet between the object and the film so that it will be radiographed together with the object?

 
 Reply 
 
Arunprakash
NDT Inspector,
Caterpillar India, India, Joined Apr 2009, 30

Arunprakash

NDT Inspector,
Caterpillar India,
India,
Joined Apr 2009
30
10:57 Jul-07-2016
Re: Edge scatter in radiography
In Reply to Dr. Yossi Shoef at 16:17 Jul-05-2016 (Opening).

Scattering reduces Signal to Noise Ratio of your radiographic image. Lead sheet recommended to mask the scattered low energy radiation.

You have to use the lead sheet at the end of section thickness change or edge of the part where your radiograph quality may be affected because of scattering.

Depends upon the material/geometry and exposure time the amount of scattering varies. You can do experiment and find out the required thickness of lead sheet as a scattered radiation filter. The thickness can be varied from 0.5 to 2 mm. In case you need 2 mm lead sheet as a masking cover you can use 2 no's of 1 mm sheet also (should be kept close and one over another). Since lead sheet is highly ductile (bending is ease with hands) and easy to cut, you can do this very easily.

All the best.

 
 Reply 
 
Arunprakash
NDT Inspector,
Caterpillar India, India, Joined Apr 2009, 30

Arunprakash

NDT Inspector,
Caterpillar India,
India,
Joined Apr 2009
30
11:39 Jul-07-2016
Re: Edge scatter in radiography
In Reply to Dr. Yossi Shoef at 16:17 Jul-05-2016 (Opening).




In continuation to your question, the lead sheet shall be placed in different place depends upon the part geometry, attached picture helps you to understand better.
 
 Reply 
 
Uwe Z.
Director,
BAM Berlin, Germany, Joined Jan 2010, 82

Uwe Z.

Director,
BAM Berlin,
Germany,
Joined Jan 2010
82
08:40 Jul-12-2016
Re: Edge scatter in radiography
In Reply to Arunprakash at 11:39 Jul-07-2016 .

Daer Arunprakash,
your sketch is nice but unfortunately not correct. We wrote this text at BAM because the lead should shield the scattered radiation out of the object just before the detector. This is always the problem when the detector is in dierct contact with the object, especially at higher energies. Your suggestion is to reduce the primary radiation BEFORE the object, but in my experinece this will generate unwanted artifacts in your image. A plain lead screen just in front of the detector will not generate artifacts to your image.
Best regards
Uwe Z.

 
 Reply 
 
Arunprakash
NDT Inspector,
Caterpillar India, India, Joined Apr 2009, 30

Arunprakash

NDT Inspector,
Caterpillar India,
India,
Joined Apr 2009
30
09:34 Jul-12-2016
Re: Edge scatter in radiography
In Reply to Uwe Z. at 08:40 Jul-12-2016 .

Thanks for highlighting the error in the picture, I was intended to keep the lead sheet close to film to filter the scatter radiation, by mistake it was shown at the top as a cover for the primary radiation. We have experimented with Ir-192 source similar geometry and got good result.
Now, I understand from your statement that entire film needs coverage of lead screen as a scatter radiation filter.

Thank again.

Arunprakash

 
 Reply 
 
Arunprakash
NDT Inspector,
Caterpillar India, India, Joined Apr 2009, 30

Arunprakash

NDT Inspector,
Caterpillar India,
India,
Joined Apr 2009
30
15:18 Jul-12-2016
Re: Edge scatter in radiography
In Reply to Uwe Z. at 08:40 Jul-12-2016 .


Back scatter filter for high thickness low thickness change region

Now I have corrected the masking position for better clarity. For the high thick zone soundness this method is suitable.

Generally the lead shall be placed between object and film cassette. (Film cassette contains intensifying lead screen, film & Back scatter filter lead screen as usual.
 
 Reply 
 

Product Spotlight

MIZ®-21C: Truly Affordable Eddy Current Handheld with Surface Array Capability

Introducing MIZ-21C, the first truly affordable handheld eddy current instrument with surface array
...
capabilities. MIZ-21C can deliver fast, accurate inspections in demanding NDT applications including aerospace, oil and gas, manufacturing, and power generation. The surface array solution can reduce inspection time by up to 95% compared to traditional handheld pencil probes. The ergonomic design, long battery life, and intuitive touchscreen mean you can inspect more areas faster than ever without fatigue.
>

Navic - Steerable Modular Automated Scanner

The Navic is a modular, motorized, steerable scanner designed to carry multiple attachments used
...
in various scanning and inspection applications. The Navic is capable of weld scanning (girth welds and long seam welds), automated corrosion mapping, and tank scanning.
>

Immersion systems

ScanMaster ultrasonic immersion systems are designed for high throughput, multi shift operation in a
...
n industrial or lab environment. These fully integrated systems provide various scanning configurations and incorporate conventional and phased arrays technologies to support diverse applications, such as inspection of disks, bars, shafts, billets and plates. All of ScanMaster immersion systems are built from high accuracy scanning frames allowing for scanning of complex parts and include a multi-channel ultrasonic instrument with exceptional performance. The systems are approved by all major manufacturers for C-scan inspection of jet engine forged discs. Together with a comprehensive set of software modules these flexible series of systems provide the customer with the best price performance solutions.
>

Magnetic X-Ray Pipeline Crawler

Zhong Yi brand pipeline crawler is magnetic/micro-video controlled crawler with DC X ray unit inst
...
alled for checking welding of pipeline. Move steadily inside the pipeline 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometers and provide the efficient inspection of the pipeline.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window