Pulse Rise Time

Recently I sent a ultrasonic thickness gage for annual calibration. It's reported that the equipment was failed on the parameter called "rise time". The Rise Time was measured to be excess 2 times of required value, but the thickness measurement accuracy result was PASS. Logically, as a user we only concern whether the measurement accuracy of the equipment is well maintain or not. Hence I would like to enquire experts here how the parameter Rise Time affect the thickness gage performance? What is the potential risk to the measurement if that Rise Time parameter not correct immediately.

Thanks all.

 
 Reply 
 
Daniel Braun
Daniel Braun
08:03 Sep-11-2016
Re: Pulse Rise Time
In Reply to Steven at 17:30 Sep-10-2016 (Opening).

Every ultrasonic instrument is an electronic device and its parameters should match with the nominal values declared by the manufacturer - the annual verification is performed in order to ensure the said matching or finding the deviations. Beside the instrument there are other factors involved into getting the instrument readings: the probe, the operator's skills, etc. The rise time is the parameter characterizing the quality of the pulser, in some cases the deviation of the said parameter from the nominal value declared by the manufacturer may affect the precision of the thickness measurements, especially for the thin materials; but sometimes - not - that was your observation: probably the actual rise time that was found for your instrument has no sensible influence while working with the probe that you used

 
 Reply 
 
Linas Svilainis
R & D,
Kaunas University of Technology, Lithuania, Joined Nov 1998, 68

Linas Svilainis

R & D,
Kaunas University of Technology,
Lithuania,
Joined Nov 1998
68
21:51 Sep-11-2016
Re: Pulse Rise Time
In Reply to Steven at 17:30 Sep-10-2016 (Opening).

Pulse rise time defines max frequency achievable by pulser:
1/pi*tR

Simply speaking, if maximum operating frequency is fmax, then rise time should be 1/3 of 1/fmax (actually it's fall time, refer Fig.11 of EN 12668-1).
Example: if max 5MHz probe is used tR should not exceed 60ns. At least.
If not, excitation signal bandwidth does not match the probe used. SNR is reduced, effective bandwidth is reduced and measurement errors increase (see Cramer-Rao lower error bound).

Actually the steeper tR, the better, but EN 12668-1 requirement is plain simple: "pulse rise time tr shall be less than the maximum value quoted in the manufacturer's technical specification". Check the device specification.

 
 Reply 
 

Product Spotlight

PAUT Probes

Typical Phased Array probes have frequencies between 1MHz and 20MHz and the number of wafers is 10
...
to 128. M2 Electronics offers customers conventionally ultrasound probes and the ability to provide high-precision Phased Array Ultrasound Probes of up to 256 wafers. We can also customize the probe for our customers to meet the specific application requirements of the user.
>

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

GEKKO - Portable Phased Array Testing with TFM in Real-Time

The portable phased array testing system GEKKO provides 64 parallel test channels. On creating testi
...
ng parameters the operator is assisted by the CIVA software. Due to its modular set-up the GEKKO instrument is suitable for operators of all skill levels.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window