Nothing is stated about the subject in ASME V, only stated that T225 requirement is not applicable.
So, I´d like to know what are the density limitations.
Any input will be appreciated.

 
 Reply 
 
Kratus
Consultant,
Petrobras Petroleo Brasileiro, Brazil, Joined Jun 2015, 3

Kratus

Consultant,
Petrobras Petroleo Brasileiro,
Brazil,
Joined Jun 2015
3
13:03 Oct-06-2017
Re: Density limitations for digital radiography
In Reply to joss at 22:59 Oct-05-2017 (Opening).

Good morning, please look at T-282.1

1
 
 Reply 
 
Rakesh Parmar
NDT Inspector,
India, Joined Dec 2014, 58

Rakesh Parmar

NDT Inspector,
India,
Joined Dec 2014
58
03:20 Oct-07-2017
Re: Density limitations for digital radiography
In Reply to joss at 22:59 Oct-05-2017 (Opening).

HI,


I think for Digital Radiography there is a grey value is given instead of Density.

 
 Reply 
 
Dr. Uwe Zscherpel
Director,
BAM Berlin, Germany, Joined Jan 2010, 81

Dr. Uwe Zscherpel

Director,
BAM Berlin,
Germany,
Joined Jan 2010
81
10:18 Oct-09-2017
Re: Density limitations for digital radiography
In Reply to Rakesh Parmar at 03:20 Oct-07-2017 .

If you are doinf Digital Radiography without using any film, then you are not able to fulfill any requirements on film density! Therefore, these requirements are translated into requirements on normailzed signal-to-noise ratio in you digital radiograph, see ISO 17636-2 for more details. Grey values doen not hlpe, because they can be scaled arbitrary. Only SNR in the image counts.
Uwe Z.

1
 
 Reply 
 
khalid
Engineering, ASNT and ISO NDT level III+ RPO
Saudi Arabia, Joined Aug 2009, 10

khalid

Engineering, ASNT and ISO NDT level III+ RPO
Saudi Arabia,
Joined Aug 2009
10
14:18 Oct-09-2017
Re: Density limitations for digital radiography
In Reply to joss at 22:59 Oct-05-2017 (Opening).

VIII-221.2 Procedure Demonstration. Demonstration of image quality indicator (IQI) image requirements of the written procedure on production or technique radiographs using phosphor imaging plate shall be considered satisfactory evidence of compliance with that procedure.

this mentioned in ASME 2010 under radiography using phosphor plate which is give at the end digital radio-graph.

1
 
 Reply 
 

Product Spotlight

Cygnus 6+ PRO Multi-Mode Ultrasonic Thickness Gauge

The Cygnus 6+ PRO thickness gauge is the most advance gauge within the Cygnus range with key featu
...
res including: comprehensive data logging; A-scan and B-scan display; manual gain control; Bluetooth connectivity; and much more. With its unique dual display and three measuring modes (Multiple-Echo, Echo-Echo and Single-Echo), this surface thickness gauge offers maximum versatility for inspections.
>

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

Magnetic X-Ray Pipeline Crawler

Zhong Yi brand pipeline crawler is magnetic/micro-video controlled crawler with DC X ray unit inst
...
alled for checking welding of pipeline. Move steadily inside the pipeline 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometers and provide the efficient inspection of the pipeline.
>

Research and Applications Development For NDT

The Research and Applications Development (RAD) group is a newly formed team within Acuren dedicat
...
ed to tackling challenging inspection problems. Our focus is the development of novel, field deployable, advanced inspection techniques for use in cases where standard NDT methods are ineffective. We don't wait for new innovations, we engineer them. From concept to commissioning.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window