Dear expert
U have one doubt regarding paut data , image attached along with
My question is why black dot is coming in sectoral setup.
Probe 5L64-A2, equipment mx2 16:128
Setup : 2 sectorial(Angle resolution 0.5 ) & one linear.
Scanning speed 40
Material P11
Please help
Thanks

 
 Reply 
 
sungeun,joo
R & D, UT
Prognosis Diagnosis Team, South Korea, Joined Jun 2013, 11

sungeun,joo

R & D, UT
Prognosis Diagnosis Team,
South Korea,
Joined Jun 2013
11
04:14 Mar-26-2018
Re: PAUT
In Reply to Sanjeev at 18:00 Mar-23-2018 (Opening).




Dear Sanjeev

I think it's better to set widen range and retest.
and if the same signal appears at the same angle with the same interval, you can suspect excess root penentration.
 
 Reply 
 
Mark
,
Zambia, Joined Jun 2016, 49

Mark

,
Zambia,
Joined Jun 2016
49
16:47 Mar-26-2018
Re: PAUT
In Reply to Sanjeev at 18:00 Mar-23-2018 (Opening).

Sanjeev
I think some pixels of display screen is dead.Save scan and image of sectorial scan by instrument on omni sd card.open the scan and image by PC.

 
 Reply 
 
Sanjeev
Sanjeev
20:30 Mar-27-2018
Re: PAUT
In Reply to Mark at 16:47 Mar-26-2018 .

Thanks for the reply
But In my equipment there are other setup file avilable , all other setup file having no problem axcept this setup
So is it possibly any other reason?
Please reply
Thank you

 
 Reply 
 
Lucian R.
NDT Inspector,
Romania, Joined Sep 2010, 6

Lucian R.

NDT Inspector,
Romania,
Joined Sep 2010
6
22:10 Mar-28-2018
Re: PAUT
In Reply to sungeun,joo at 04:14 Mar-26-2018 .

Dear Joo Sungeon, please let me ask you a question for your statement "if the same signal appears at the same angle with the same interval, you can suspect excess root penentration.": how can make sure (that is not only to suspect) root penetration?
Would you recommend a few steps on this matter (ensuring signal is from root)? Like - checking from the other side (is this required or only recommended?). Other tips.
Thanks,
Lucian R.

 
 Reply 
 
Nicholas Bublitz
Other,
VeriPhase, USA, Joined May 2010, 134

Nicholas Bublitz

Other,
VeriPhase,
USA,
Joined May 2010
134
23:07 Mar-28-2018
Re: PAUT
In Reply to Sanjeev at 18:00 Mar-23-2018 (Opening).

zoom image



Sanjeev,

This looks mainly like electrical noise. If the equipment was plugged in check the line for fluctuations, otherwise welding or fork truck in area etc. This could also be a bad encoder- check all cables for rips etc.

Other than that check PRF, you have higher resolution and fast scan speed and may need to adjust PRF until you do not see the noise.
 
 Reply 
 
Sanjeev
Sanjeev
17:23 Mar-29-2018
Re: PAUT
In Reply to Nicholas Bublitz at 23:07 Mar-28-2018 .

Thank you Nicholas sir
I think you r right. I will try ,
Thank you very much

 
 Reply 
 
Sungeun, Joo
R & D, UT
Prognosis Diagnosis Team, South Korea, Joined Jun 2013, 11

Sungeun, Joo

R & D, UT
Prognosis Diagnosis Team,
South Korea,
Joined Jun 2013
11
08:57 Apr-02-2018
Re: PAUT
In Reply to Sanjeev at 17:23 Mar-29-2018 .

Sanjeev

I misunderstood your question.
'Black dot' should be Electrorical noise. somtimes this happen when somebody weld around(cause of welding earth).

Mr.Lucian.R
Please mail me what you want to know.

 
 Reply 
 

Product Spotlight

Ultrasonic Squirter/Gantry System

TecScan’s NDT Ultrasonic Gantry Systems are industrial Squirter scanners designed for non-destru
...
ctive quality testing and raster scanning of large structures and parts. The Gantry Systems are usually composed of an Industrial Mechanical Scanner and a Control Room with System Workstation.
>

IRIS 9000Plus - Introducing the next generation of heat exchanger inspection.

Representing the seventh generation of the IRIS system, the IRIS 9000 Plus has nearly 200 years of c
...
ombined field inspection experience incorporated in its design. This experience combined with a strong commitment to quality and a history of innovation has made Iris Inspection Services® the undisputed leader in IRIS technology.
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

New PHAsis module for longitudinal weld seam inspection

PHAsis® Phased Array manual ultrasonic spot weld inspection device is now ideally suited for test
...
ing of longitudinally welded seams (e.g. Arplas® seam, short line seams) with a length of 3 to 12 mm.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window