10:17 Jun-20-2016
Dead Zone calculation

There is always a minimum thickness for any particular UT setup where the results are not sensitive,which is behalf of the interference effect close to the transducer,which we call as dead zone
How to calculate the dead zone

If somewhere i am wrong in my concept please rectify me

Thanks to you all in advance

 
 Reply 
 
Anmol Birring
Consultant,
Birring NDE Center, Inc., USA, Joined Aug 2011, 785

Anmol Birring

Consultant,
Birring NDE Center, Inc.,
USA,
Joined Aug 2011
785
11:42 Jun-20-2016
Re: Dead Zone calculation
In Reply to Dhiliban at 10:17 Jun-20-2016 (Opening).

There is no calculation. You just have to measure it.

 
 Reply 
 
massimo carminati
Consultant, AUT specialist
IMG Ultrasuoni Srl, Italy, Joined Apr 2007, 691

massimo carminati

Consultant, AUT specialist
IMG Ultrasuoni Srl,
Italy,
Joined Apr 2007
691
23:08 Jun-20-2016
Re: Dead Zone calculation
In Reply to Dhiliban at 10:17 Jun-20-2016 (Opening).

The dead zone of a UT measuring chain and test condition has to be measured and is the minimum distance from beam entry point where an echo can be distinguished.

1
 
 Reply 
 
Satyajeet Jena
Satyajeet Jena
08:54 Aug-03-2018
Re: Dead Zone calculation
In Reply to massimo carminati at 23:08 Jun-20-2016 .

It is a short time at which the transmitted signal blinds the received signal due to high noise of the crystal. Just like eye blinking. Happens in probe which transmit & receive in one crystal..

Thanks

 
 Reply 
 
langtuteng
Armenia, Joined Nov 2014, 196

langtuteng

Armenia,
Joined Nov 2014
196
03:45 Aug-05-2018
Re: Dead Zone calculation
In Reply to Dhiliban at 10:17 Jun-20-2016 (Opening).

Yes, the dead zone is measured. it is based on your probe, equipment and sensitivity. and it varies from different sensitivity. so I suggest you to measure it under the probe and equipment combined system, also with the sensitivity. the dead zone is the distance where you can distinguish the first echo behined the initial pulse echo, and mostly we define if the trough between the initial pulse wave and first echo wave is 6dB lower than the peak of first echo wave , then we can distiguish them and this diatance is dead zone.

2
 
 Reply 
 
Arman
Engineering,
Turkey, Joined Oct 2014, 39

Arman

Engineering,
Turkey,
Joined Oct 2014
39
08:48 Aug-05-2018
Re: Dead Zone calculation
In Reply to massimo carminati at 23:08 Jun-20-2016 .

Gents,

Thanks for all explanations, could you please explain effect of parameters like probe frequency, crystal diameter ...etc on length of deadzone?

 
 Reply 
 
Paul Holloway
Consultant,
Holloway NDT & Engineering Inc , Canada, Joined Apr 2010, 227

Paul Holloway

Consultant,
Holloway NDT & Engineering Inc ,
Canada,
Joined Apr 2010
227
10:39 Aug-05-2018
Re: Dead Zone calculation
In Reply to Arman at 08:48 Aug-05-2018 .

There was a lengthy discussion regarding dead zone here:

https://www.ndt.net/forum/thread.php?msgID=62184

A simple forum orGoogle search would have revealed this.

1
 
 Reply 
 

Product Spotlight

EKOSCAN Phased Array

In order to always fit your needs, EKOSCAN can manufacture any type of UT transducer, either convent
...
ional or Phased Array. As an ISO 9001: 2015 certified company, EKOSCAN is extremely careful as far a material selection and manufacturing processes are concerned. Our probes guarantee our customers the benefits of latest innovations regarding piezo-composite, backing, impedance adaptation layer, etc. Specific probes for hostile environment: high temperature, high pressure, corrosive environment,etc. Specific probes designed to fit your specific application: optimization of every parameter to guarantee you the best detection.
>

NEOS III

NEOS III is Logos Imagings lightest DR system. With a built-in battery and internal wireless commu
...
nication, the NEOS III is perfect for users that want to quickly assess an item.
>

NovaScope 6000

The all-digital Novascope 6000 is a portable, ultra-high precision thickness gauge for high-speed
...
thickness measurement. Novascope 6000 has unmatched capabilities and unique features including: •Superior Resolution with high contrast, high-speed color RF display •High pulser voltage •Real-time video output •Increased internal/external data storage •Programmable SetUp features •Battery & AC Powered
>

Combination of Digital Image Correlation and Thermographic Measurements