Echo on A Scan not displaying
zoom image

C SCAN IMAGE

Good Day fellow professionals,

Please can somebody share their experience with the below problem and how it was resolve. SEE ATTACHED IMAGE

During setup/calibration for corrosion mapping using Olympus MX2 32:128PR with Hydroform, the echo on the A-scan does not show but all the B and C-scan are showing, I have changed scanner but the problem still exist. What could be the problem? Is the problem with the scanners or the equipment?

Thanks to assist

 
 Reply 
 
Ali
NDT Inspector,
Iran, Joined Mar 2013, 117

Ali

NDT Inspector,
Iran,
Joined Mar 2013
117
15:13 Feb-08-2020
Re: Echo on A Scan not displaying
In Reply to Michael Eneojo at 13:18 Feb-08-2020 (Opening).

Michael,
I think the A-Scan data is linked only to C- scan index axis blue curser , not to B scan yellow curser (there must be another blue curser on B-Scan that is linked to A &C scan).would you please apply zooming on C-scan and move index and scan axis cursers on the blue colour area and capture another photo to upload here. Also some elements of probe are off.

 
 Reply 
 
Paul Holloway
Consultant,
Holloway NDT & Engineering Inc , Canada, Joined Apr 2010, 242

Paul Holloway

Consultant,
Holloway NDT & Engineering Inc ,
Canada,
Joined Apr 2010
242
15:36 Feb-08-2020
Re: Echo on A Scan not displaying
In Reply to Michael Eneojo at 13:18 Feb-08-2020 (Opening).

zoom image

Set VPA to something else

As Ali has pointed out, set your VPA to 30. That will put it in the middle of your probe. See image.
 
 Reply 
 
Paul Holloway
Consultant,
Holloway NDT & Engineering Inc , Canada, Joined Apr 2010, 242

Paul Holloway

Consultant,
Holloway NDT & Engineering Inc ,
Canada,
Joined Apr 2010
242
02:19 Feb-10-2020
Re: Echo on A Scan not displaying
In Reply to Michael Eneojo at 13:18 Feb-08-2020 (Opening).

And you've got WAY too much range on there. Turn it down to 30mm, and put the start at -3mm (or something just a little less than zero). I can see your nominal is 12.7mm.

Then you can see two backwalls and play with cursor positions to remove coating if the E-E signal is inconsistent.

 
 Reply 
 
ANDY
,
China, Joined Mar 2012, 20

ANDY

,
China,
Joined Mar 2012
20
18:54 Feb-13-2020
Re: Echo on A Scan not displaying
In Reply to Michael Eneojo at 13:18 Feb-08-2020 (Opening).

Please try to reconnect your probe again. It seems like your probe connector is not connect tight

 
 Reply 
 

Product Spotlight

Micro CT

nanoVoxel series has nano-scale resolution by using high voltage X-ray source (from 80-300 kV) wit
...
h micro focal spot and highly sensitive detector including different kinds of big size panel detectors and lens coupling CCD.
>

New PHAsis module for longitudinal weld seam inspection

PHAsis® Phased Array manual ultrasonic spot weld inspection device is now ideally suited for test
...
ing of longitudinally welded seams (e.g. Arplas® seam, short line seams) with a length of 3 to 12 mm.
>

HD-CR 35 NDT Computed Radiography System

Portable high-resolution CR scanner for all radiography applications - weld testing, profile images
...
and aerospace. No matter what type of radiographic testing you are performing, the unique TreFoc Technology of the HD-CR 35 NDT imaging plate scanner always guarantees the highest image quality.
>

Semi-Automated Phased Array Immersion System for Small Composite Parts

Turn-key semi-automated system as an improved and affordable solution for inspection of small comp
...
osite parts. Includes support table, immersion tank, scanner, PA instrument, PC, Analysis software, database, wedge management and other options.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window