09:25 Nov-02-2006
TCG CURVE

Dear friends,
Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
Depth of the material: 74mm
Reference reflector: 1.5mm FBH 15mm depth

Your reply will be very useful for me.


Best Regards,


Debdutta Mallik


 
 Reply 
 
Linxiao
Linxiao
08:37 Nov-06-2006
Re: TCG CURVE
Following the link (session 33), you can find several papers on the "2004 ultrasonic benchmark" problem. Their calculations can answer your questions.

good luck!


----------- Start Original Message -----------
: Dear friends,
: Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
: Depth of the material: 74mm
: Reference reflector: 1.5mm FBH 15mm depth
: Your reply will be very useful for me.
:
: Best Regards,
:
: Debdutta Mallik
------------ End Original Message ------------




 
 Reply 
 
Linxiao
Linxiao
08:37 Nov-06-2006
Re: TCG CURVE
Following the link (session 33), you can find several papers on the "2004 ultrasonic benchmark" problem. Their calculations can answer your questions.

http://www.cnde.iastate.edu/QNDE/2004%20Info/Abstract%20Book.pdf

good luck!


----------- Start Original Message -----------
: Dear friends,
: Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
: Depth of the material: 74mm
: Reference reflector: 1.5mm FBH 15mm depth
: Your reply will be very useful for me.
:
: Best Regards,
:
: Debdutta Mallik
------------ End Original Message ------------




 
 Reply 
 
Joe Buckley
Consultant, ASNT L-III, Honorary Secretary of BINDT
Level X NDT, BINDT, United Kingdom, Joined Oct 1999, 523

Joe Buckley

Consultant, ASNT L-III, Honorary Secretary of BINDT
Level X NDT, BINDT,
United Kingdom,
Joined Oct 1999
523
09:44 Nov-06-2006
Re: TCG CURVE
You cannot construct a curve from one point...

However if you have a very good theoretical understanding of the system it might be possible to calculate a curve (in a way similar to the DGS system.) the single reference then allows a sensitivity alignment.

The risk of this is that if you make any incorrect assumptions the curve wont be valid.

Using three or more reference points is the only reliable experimental approach.

Joe


----------- Start Original Message -----------
: Dear friends,
: Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
: Depth of the material: 74mm
: Reference reflector: 1.5mm FBH 15mm depth
: Your reply will be very useful for me.
:
: Best Regards,
:
: Debdutta Mallik
------------ End Original Message ------------




 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1274

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1274
02:03 Nov-07-2006
Re: TCG CURVE
If the probe you are using is circular the equations derived by Prof. I. N. Ermolov may be useful to derive an approximate curve. I provided the solutions in a user friendly form in an NDT.net article a few years ago.
http://www.ndt.net/article/v07n01/ginzel/ginzel.htm

By entering the attenuation as 0dB/mm you will get the uncorrected version of the relative amplitudes of the selected reflector. i.e. this is the approximate effect of just the spherical divergence effects.

Ed


----------- Start Original Message -----------
: You cannot construct a curve from one point...
: However if you have a very good theoretical understanding of the system it might be possible to calculate a curve (in a way similar to the DGS system.) the single reference then allows a sensitivity alignment.
: The risk of this is that if you make any incorrect assumptions the curve wont be valid.
: Using three or more reference points is the only reliable experimental approach.
: Joe
:
: : Dear friends,
: : Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
: : Depth of the material: 74mm
: : Reference reflector: 1.5mm FBH 15mm depth
: : Your reply will be very useful for me.
: :
: : Best Regards,
: :
: : Debdutta Mallik
------------ End Original Message ------------




 
 Reply 
 
Khurath
SIS, India, Joined Jan 2018, 2

Khurath

SIS,
India,
Joined Jan 2018
2
06:25 Mar-14-2019
Re: TCG CURVE
In Reply to Joe Buckley at 09:44 Nov-06-2006 .

Thanks for this information of sensitivity calibration and for above 74mm thickness what should be our material debt for calibration

 
 Reply 
 

Product Spotlight

Teletest Focus+

Teletest Focus+ electronics have superior capabilities than rivals on the market. Beyond the usual
...
test features, Focus+ has 24 transmit channels and 24 receive channels, with an additional on-board switching capabilities. The instrument's frequency range is 10–300 kHz.
>

Magnetic X-Ray Pipeline Crawler

Zhong Yi brand pipeline crawler is magnetic/micro-video controlled crawler with DC X ray unit inst
...
alled for checking welding of pipeline. Move steadily inside the pipeline 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometers and provide the efficient inspection of the pipeline.
>

NovaScope 6000

The all-digital Novascope 6000 is a portable, ultra-high precision thickness gauge for high-speed
...
thickness measurement. Novascope 6000 has unmatched capabilities and unique features including: •Superior Resolution with high contrast, high-speed color RF display •High pulser voltage •Real-time video output •Increased internal/external data storage •Programmable SetUp features •Battery & AC Powered
>

YXLON Cougar EVO

Scalable small footprint X-ray inspection systems for assembly and laboratory applications. The
...
YXLON Cougar EVO series was designed to provide the "best-in- class" inspection solutions for SMT, semiconductor, and laboratory assembly applications, while maintaining a small system footprint for maximum convenience. With optimized software and hardware, these systems produce higher quality and more consistent results than other electronics inspection systems currently on the market.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window