06:33 Feb-25-2007
PA Gap Dimensions

Can someone explain to me exactly what the gap dimension is for a phased array probe? I haven't seen it on any probe data sheets but I have heard a few different explantions. Does it relate to the pitch?

I need to know as it is a recording requirement of ASME Sec. V, Art. 4, Appendix 4 - Examination Record (a).

Thanks,

Joel Norman


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1303

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1303
01:00 Feb-25-2007
Re: PA Gap Dimensions
Joel:
The gap is just the space between the elements (i.e. the thickness of the saw that was used to slice the array). If you have a copy of the RD Tech "Introduction to Phased Array Ultrasonic Technology Applications" book, it is shown on page 107.
The Probe parameters such as these are essential variables so need to be itemised for a proper record of the equipment used.

Regards
Ed

----------- Start Original Message -----------
: Can someone explain to me exactly what the gap dimension is for a phased array probe? I haven't seen it on any probe data sheets but I have heard a few different explantions. Does it relate to the pitch?
: I need to know as it is a recording requirement of ASME Sec. V, Art. 4, Appendix 4 - Examination Record (a).
: Thanks,
: Joel Norman
------------ End Original Message ------------




 
 Reply 
 
Joel Norman
NDT Inspector
Canada, Joined Jun 2008, 5

Joel Norman

NDT Inspector
Canada,
Joined Jun 2008
5
03:55 Feb-26-2007
Re: PA Gap Dimensions
----------- Start Original Message -----------
: Joel:
: The gap is just the space between the elements (i.e. the thickness of the saw that was used to slice the array). If you have a copy of the RD Tech "Introduction to Phased Array Ultrasonic Technology Applications" book, it is shown on page 107.
: The Probe parameters such as these are essential variables so need to be itemised for a proper record of the equipment used.
: Regards
: Ed
:
: : Can someone explain to me exactly what the gap dimension is for a phased array probe? I haven't seen it on any probe data sheets but I have heard a few different explantions. Does it relate to the pitch?
: : I need to know as it is a recording requirement of ASME Sec. V, Art. 4, Appendix 4 - Examination Record (a).
: : Thanks,
: : Joel Norman
------------ End Original Message ------------

Thanks Ed.


 
 Reply 
 

Product Spotlight

Extended Range Variable Wall® Piping Calibration Blocks for Longseam Inspection

This set of two blocks has been specially designed for situations where ultrasonic examination tak
...
es place circumferentially or “around the curve” for applications such as inspection of longseam welds in piping and vessels. Jointly designed by PH Tool and Holloway NDT, the blocks feature side-drilled hole reflectors positioned within novel compound-curves, each with multiple thickness steps. The unique design results in a light and compact form factor, with the largest block weighing just over 15 lbs and spanning just 16 ¾”. Coverage is provided for nominal pipe sizes from 6” to 10” in the small block and 12” to a whopping 56” in the large block. The compound OD curves and multiple wall thicknesses actually exceed ASME requirements, providing exceptional signal response and measurement accuracy. Properly managing beam divergence off the ID and stabilizing beam skew due to probe rocking is essential to maintaining a consistent calibration when inspecting lengthwise on curved surfaces.
>

Conformable wedge transducer

The conformability is obtained with a flexible membrane filled with water between the transducer and
...
the inspected component. The coupling between the membrane and the component requires a small quantity of water or couplant. The conformable wedge combines the acoustic performance of immersion technique with good coupling and low attenuation.
>

Micro CT

nanoVoxel series has nano-scale resolution upto 500nm by using high voltage X-ray source (from 80-
...
300 kV) with micro/nano focal spot and highly sensitive flat panel detectors and unique Optical coupling detector.
>

Ultrasonic Probe Recharacterization Service

NDT Systems offers a comprehensive Aftercare and Recharacterization Service for all our ultrasonic
...
probes. The Recharacterization Service is fully compliant with International ASTM E1065 Standard Guide (and other applicable standards) and offers complete documentation, traceable to the ASTM E1065 Standard. For more details or to schedule Recharacterization Services contact ndtsales@ndtsystems.com
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window