coverage for TOFD

I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1286

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1286
03:11 Jul-25-2007
Re: coverage for TOFD
----------- Start Original Message -----------
: I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.
------------ End Original Message ------------
The "missed data" would be the dead zone. This is not actually a function of the wall thickness but instead, the ring time equivalent depth of the pulse. Frequency, probe spacing and angle used can all be factors. A simple calculator is provided as a download on NDT.net to estimate the Dead Zone heights associated with TOFD setups. Go to http://www.ndt.net/article/v10n06/ginzel/ginzel.htm
For it to be 1/3 wall thickness the specimen you are working with is probably very thin.


 
 Reply 
 
sentho
sentho
05:54 Jul-29-2007
Re: coverage for TOFD
- The "missed data" would be the dead zone.-his is from your replay. How it happents? We are using one probe as transmitter and the other is a receiver.--------- Start Original Message -----------
: : I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.
: The "missed data" would be the dead zone. This is not actually a function of the wall thickness but instead, the ring time equivalent depth of the pulse. Frequency, probe spacing and angle used can all be factors. A simple calculator is provided as a download on NDT.net to estimate the Dead Zone heights associated with TOFD setups. Go to http://www.ndt.net/article/v10n06/ginzel/ginzel.htm
: For it to be 1/3 wall thickness the specimen you are working with is probably very thin.
------------ End Original Message ------------




 
 Reply 
 
Nigel Armstrong
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel Armstrong

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
02:31 Jul-29-2007
Re: coverage for TOFD
Sentho

The dead zone which Ed is refering to is not of the same origin as the dead zone with the single crystal pulse-echo technique which is a characteristic of the transducer vibration. The TOFD dead zone is caused by the surface or lateral wave travelling from the transmitting to the receiving probe. Any near-surface diffracted signals will be swamped by the lateral wave.

Regards

Nigel


---------- Start Original Message -----------
: - The "missed data" would be the dead zone.-his is from your replay. How it happents? We are using one probe as transmitter and the other is a receiver.--------- Start Original Message -----------
: : : I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.
: : The "missed data" would be the dead zone. This is not actually a function of the wall thickness but instead, the ring time equivalent depth of the pulse. Frequency, probe spacing andangle used can all be factors. A simple calculator is provided as a download on NDT.net to estimate the Dead Zone heights associated with TOFD setups. Go to http://www.ndt.net/article/v10n06/ginzel/ginzel.htm
: : For it to be 1/3 wall thickness the specimen you are working with is probably very thin.
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

ISONIC 3510

Very Powerful Superior Performance Extremely Portable Smart Phased Array Ultrasonic Flaw Detector an
...
d Recorder with 2 Conventional UT and TOFD Channels uniquely combines PA, single- and multi-channel conventional UT, and TOFD modalities providing 100% raw data recording and imaging. Suitable for all kinds of every-day ultrasonic inspections
>

FD800 Bench Top Flaw Detectors

The bench-top FD800 flaw detector range combines state-of-the-art flaw detection with advanced mater
...
ial thickness capabilities. Designed for use in the laboratory these gauges are the tool you need for all your flaw detecting needs.
>

Varex Imaging Large Field of View (FOV) Digital Detector Arrays (DDAs)

A larger FOV DDA can reduce the space and volume of the X-ray inspection system on the factory floor
...
, enable faster scanning times, better throughput and better resolution images at a lower dose. Customers can also save time and money. With these benefits in mind, Varex Imaging has designed a family of large FOV detectors (4343HE, XRD 1611, 4343DX-I, 4343CT) for our industrial imaging customers.
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window