sentho
sentho
01:30 Jul-25-2007
coverage for TOFD

I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.


    
 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1268

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1268
03:11 Jul-25-2007
Re: coverage for TOFD
----------- Start Original Message -----------
: I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.
------------ End Original Message ------------
The "missed data" would be the dead zone. This is not actually a function of the wall thickness but instead, the ring time equivalent depth of the pulse. Frequency, probe spacing and angle used can all be factors. A simple calculator is provided as a download on NDT.net to estimate the Dead Zone heights associated with TOFD setups. Go to http://www.ndt.net/article/v10n06/ginzel/ginzel.htm
For it to be 1/3 wall thickness the specimen you are working with is probably very thin.


    
 
 Reply 
 
sentho
sentho
05:54 Jul-29-2007
Re: coverage for TOFD
- The "missed data" would be the dead zone.-his is from your replay. How it happents? We are using one probe as transmitter and the other is a receiver.--------- Start Original Message -----------
: : I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.
: The "missed data" would be the dead zone. This is not actually a function of the wall thickness but instead, the ring time equivalent depth of the pulse. Frequency, probe spacing and angle used can all be factors. A simple calculator is provided as a download on NDT.net to estimate the Dead Zone heights associated with TOFD setups. Go to http://www.ndt.net/article/v10n06/ginzel/ginzel.htm
: For it to be 1/3 wall thickness the specimen you are working with is probably very thin.
------------ End Original Message ------------




    
 
 Reply 
 
Nigel Armstrong
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel Armstrong

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
02:31 Jul-29-2007
Re: coverage for TOFD
Sentho

The dead zone which Ed is refering to is not of the same origin as the dead zone with the single crystal pulse-echo technique which is a characteristic of the transducer vibration. The TOFD dead zone is caused by the surface or lateral wave travelling from the transmitting to the receiving probe. Any near-surface diffracted signals will be swamped by the lateral wave.

Regards

Nigel


---------- Start Original Message -----------
: - The "missed data" would be the dead zone.-his is from your replay. How it happents? We are using one probe as transmitter and the other is a receiver.--------- Start Original Message -----------
: : : I noticed that every TOFD inspection miss some datas from the upper wall thickness(approximately 1/3 T). how can we get a better quality signals from the laterwave side.
: : The "missed data" would be the dead zone. This is not actually a function of the wall thickness but instead, the ring time equivalent depth of the pulse. Frequency, probe spacing andangle used can all be factors. A simple calculator is provided as a download on NDT.net to estimate the Dead Zone heights associated with TOFD setups. Go to http://www.ndt.net/article/v10n06/ginzel/ginzel.htm
: : For it to be 1/3 wall thickness the specimen you are working with is probably very thin.
------------ End Original Message ------------




    
 
 Reply 
 

Product Spotlight

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

AMIGO2

TSC Amigo2 - ACFM technology has developed a solid reputation for accurately detecting and sizing
...
surface-breaking cracks through paint and coatings. As the industry demands increased performance in speed, signal quality, and portability, it’s time for an evolution. It’s time for Amigo2.
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

Combination of Digital Image Correlation and Thermographic Measurements

The combination of measuring results from the digital image correlation (ARAMIS, DIC) and temperat
...
ure measuring data from infrared cameras permits the simultaneous analysis of the thermal and mechanical behavior of test specimens in the materials and components testing field.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window