03:58 Nov-19-1999
beam width of focusing probe (Antonio Filippi)

reply by Udo Schlengermann
uschlengermann@krautkramer.de

The parameters you gave are valid for an unfocused flat probe.
Assuming a (1/4) inch probe of 10 MHz in water (6,35 mm dia)
this probe has its natural focus at 68 mm distance (nearfield length) and a beam width of 1,6 mm at a -6dB-threshold (measured with a point like reflector).

The 9 mm steel path is equivalent to 36 mm water path, so you have to use a water delay of 32 mm (68 mm -36 mm) to position the focus at the far surface of the steel plate.

This will not influence the beam width, if the beam is perpendicular to the plate.

Kind regards

Udo Schlengermann




 
 Reply 
 

Product Spotlight

TraiNDE UT

TraiNDE UT is a virtual tool associated with a signal database which simulates real inspection con
...
ditions for numerous applications (Type A/V1 block, DAC block, welds and plates).
>

NEOS III

NEOS III is Logos Imagings lightest DR system. With a built-in battery and internal wireless commu
...
nication, the NEOS III is perfect for users that want to quickly assess an item.
>

Surf-X® Array Probe

Introducing the Surf-X family of flexible Eddy Current Array (ECA) probes. Featuring unique multiple
...
coil sets and proprietary X-PROBE technology, Surf-X array probes can quickly and accurately handle a range of inspection applications, from inspecting corrosion or cracking in pipes, pressure vessels, or tanks, to assessing and sizing cracks in raised welds and friction stir welds.
>

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window