Azzam
Azzam
08:55 Sep-07-2007
IQI/penetrameter for Exotic Material?

Hello guys..
would appreciate if u guys can assist me in determining the suitable IQI for RT on exotic material such as Hastelloy-X, Nimonic 263 and others. Is there any IQI made of these materials? can i just simply use the IQI that we use on carbon steel? Thanking u in advance!


 
 Reply 
 
DJ Kallhof
DJ Kallhof
03:25 Sep-07-2007
Re: IQI/penetrameter for Exotic Material?
----------- Start Original Message -----------
: Hello guys..
: would appreciate if u guys can assist me in determining the suitable IQI for RT on exotic material such as Hastelloy-X, Nimonic 263 and others. Is there any IQI made of these materials? can i just simply use the IQI that we use on carbon steel? Thanking u in advance!
------------ End Original Message ------------

Both Hastelloy-X and Nimonic 263 are primarily Nickel based alloys. Without knowing the code you are working to I am hesitant to state wether you are able to use anything else. However, I believe you are able to substitute a softer material, ie. Steel for Nickel, when you are in a pinch. The down side is, it will require thicker shimming of the steel to compensate for the hardness difference. Remember to select the IQI based on the material thickness you are attempting to radiograph, not on the thickness of the shim you are required to use to even out the densities.


 
 Reply 
 
ET
ET
04:19 Sep-12-2007
Re: IQI/penetrameter for Exotic Material?
----------- Start Original Message -----------
: : Hello guys..
: : would appreciate if u guys can assist me in determining the suitable IQI for RT on exotic material such as Hastelloy-X, Nimonic 263 and others. Is there any IQI made of these materials? can i just simply use the IQI that we use on carbon steel? Thanking u in advance!
: Both Hastelloy-X and Nimonic 263 are primarily Nickel based alloys. Without knowing the code you are working to I am hesitant to state wether you are able to use anything else. However, I believe you are able to substitute a softer material, ie. Steel for Nickel, when you are in a pinch. The down side is, it will require thicker shimming of the steel to compensate for the hardness difference. Remember to select the IQI based on the material thickness you are attempting to radiograph, not on the thickness of the shim you are required to use to even out the densities.
------------ End Original Message ------------

That is correct. You can use an IQI from a lower group number as defined in ASME Sec. V, basically less dense materials



 
 Reply 
 
Vishal Gupta
Vishal Gupta
15:18 Jun-16-2013
Re: IQI/penetrameter for Exotic Material?
In Reply to DJ Kallhof at 03:25 Sep-07-2007 .

For Selection of the penetrameters you can refer ASTM E-747. For Nickel based alloys contaning Ni-Cr-Mo as major elements and Ni-Cu alloys use Group-3 and group-4 penetrameter respectively.

 
 Reply 
 

Product Spotlight

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

NEOS III

NEOS III is Logos Imagings lightest DR system. With a built-in battery and internal wireless commu
...
nication, the NEOS III is perfect for users that want to quickly assess an item.
>

NDT Master Lecturer

In the program both university professors and practitioners will give lectures, which guarantees the
...
oretical depth and practical inside. Academic Director: Prof. Dr. Christian Boller
The following lecturers are not complete: Prof. Tadeusz Stepinski, Prof. Wieslaw Staszewski, Prof. Frank Walther, Prof. Giovanni Bruno, Prof. Gerd Dobmann, Prof. Philippe Guy
>

PROlineTOP Plug & Play Ultrasonic inspection device

As Plug & Play solution it units all control and operation elements in a small housing and therefo
...
re replaces the typical control cabinet...
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window