Transducer failures

My company has experienced a recent spate of ultrasonic transducer failures due to blistering of the epoxy facing of cylindrically focused immersion transducers, 3 in the last two months. One transducer was not quite one year old. The other two were approximately 5 years old. Does anyone know of a possible cause of transducer blistering? Perhaps an additive or change in chemical composition of the water? Are there any known chemicals which could attack the epoxy faces of transducers? Could sodium sulfite be a cause?


 
 Reply 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
08:16 Oct-17-2007
Re: Transducer failures
Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.

In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1303

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1303
03:06 Oct-18-2007
Re: Transducer failures
I would like to first concur with Tom and then just add that if the water temperature has been elevated (e.g. 25-35C) it may accelerate the situation

----------- Start Original Message -----------
: Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.
: In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

Cygnus 6+ PRO Multi-Mode Ultrasonic Thickness Gauge

The Cygnus 6+ PRO thickness gauge is the most advance gauge within the Cygnus range with key featu
...
res including: comprehensive data logging; A-scan and B-scan display; manual gain control; Bluetooth connectivity; and much more. With its unique dual display and three measuring modes (Multiple-Echo, Echo-Echo and Single-Echo), this surface thickness gauge offers maximum versatility for inspections.
>

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

SITEX CPSERIES

Teledyne ICM’s CPSERIES has been designed with a view to revolutionizing the handling and perfor
...
mances of portable X-Ray sets. Despite having managed to halve the weight of similar portable X-Ray generators available on the market (while continuing to provide the same power output), the SITEX CPSERIES generators feature a shutter, a laser pointer, a beryllium window, an aluminum filter and two integrated diaphragms (customized sizes are available upon request). Without compromising the robustness and reliability for which ICM products are renowned, the small size and light weight of the SITEX CPSERIES will radically change the way that you perform your RT inspections. And you will see a positive impact in terms of both quality and return on investment (ROI).
>

OPBOX with standard software is able to do all types of inspections and measurements: flaw detection in welds and materials, scanning of objects, testing composite materials, forged and moulded pieces, many UT inspections, measurements of properties of ma

Typical applications: UT measurements with pulse technique, Measurement of thicknesses also at hig
...
h temperatures, Measurements of properties of materials, including fluids and gases We are delivering a standard version of the software (for any Microsoft Windows up to 10 x64 with Microsoft Hardware certification report Approved) and for special needs: SDK with ready to use examples for LabView, MATLAB x64, C++ wrapper for dll, Python and Linux., and also low-level description of how to control our devices directly from any USB tools
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window