My company has experienced a recent spate of ultrasonic transducer failures due to blistering of the epoxy facing of cylindrically focused immersion transducers, 3 in the last two months. One transducer was not quite one year old. The other two were approximately 5 years old. Does anyone know of a possible cause of transducer blistering? Perhaps an additive or change in chemical composition of the water? Are there any known chemicals which could attack the epoxy faces of transducers? Could sodium sulfite be a cause?


 
 Reply 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
08:16 Oct-17-2007
Re: Transducer failures
Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.

In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1304

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1304
03:06 Oct-18-2007
Re: Transducer failures
I would like to first concur with Tom and then just add that if the water temperature has been elevated (e.g. 25-35C) it may accelerate the situation

----------- Start Original Message -----------
: Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.
: In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

Research and Application Development For NDT

Acuren’s Research and Application Development specializes in the development of advanced ultraso
...
nic inspection techniques and systems for challenging inspection applications, with an emphasis on practical solutions which are field deployable. Services include manual and automated ultrasonic inspection system development, inspection technique optimization using laboratory scale studies and ultrasonic modeling (CIVA, BeamTool), preparing technical justification for technique evaluation and qualification (Probability of Detection and sizing accuracy studies), inspection/calibration/analysis procedure preparation to support field deployment of custom techniques, and development of custom imaging algorithms to support challenging inspection applications.
>

32:128PR PAUT & 2-ch TOFD Flaw Detector: SyncScan 2

●32:128PR PAUT with optional 2-ch TOFD to maximize your efficiency ●Support PR mode for corrosi
...
on inspection ●Multiple solutions for welds & corrosion ●Faster scan speed (Approximately 3 meters/minute) ●Light weight: 4kg only including battery
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

FAAST-PA! OEM Patented phased Array for high speed UT inspection

Multiangle, Multifocus, Multifrequency, Multibeam. Instead of stacking UT electronics and having m
...
any PA probes, FAAST-PA is able to transmit all delay laws within ONE single shot in Real time.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window