Transducer failures

My company has experienced a recent spate of ultrasonic transducer failures due to blistering of the epoxy facing of cylindrically focused immersion transducers, 3 in the last two months. One transducer was not quite one year old. The other two were approximately 5 years old. Does anyone know of a possible cause of transducer blistering? Perhaps an additive or change in chemical composition of the water? Are there any known chemicals which could attack the epoxy faces of transducers? Could sodium sulfite be a cause?


 
 Reply 
 
Tom Nelligan
Engineering,
retired, USA, Joined Nov 1998, 390

Tom Nelligan

Engineering,
retired,
USA,
Joined Nov 1998
390
08:16 Oct-17-2007
Re: Transducer failures
Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.

In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1303

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1303
03:06 Oct-18-2007
Re: Transducer failures
I would like to first concur with Tom and then just add that if the water temperature has been elevated (e.g. 25-35C) it may accelerate the situation

----------- Start Original Message -----------
: Chemicals in your water could be a possibility, but are you soaking the transducers continuously? Epoxy blistering can happen even in distilled water over a long period of time if transducers aren't given a chance to dry out, due to water absorption in the lens epoxy which causes it to swell and blister. If you' were soaking two of them more or less continuously for five years, I'd expect it.
: In the case of Panametrics-NDT immersion transducers, we publish a recommendation that they should not be submerged for more than 8 hours at a time, followed by 16 hours of drying. That's a VERY conservative recommendation that will usually insure many years of service, and many of our customers are happy with a more liberal duty cycle, but the point is that epoxy-faced immersion transducers can indeed suffer blistering in pure water if they're soaked long enough without a drying period.
------------ End Original Message ------------




 
 Reply 
 

Product Spotlight

Echomac® PA Bar Tester

The Echomac® PA BT provides 100% inspection of round bars for core and surface defects, at speeds u
...
p to 2 m/sec, depending on the specific application. The unit allows producers to handle a much larger range of diameters in one installation, and is available in 3 models to test 10 – 75mm diameter (small model); 20 – 130mm (medium model) or 50 – 254mm (large model) bars.
>

Extended Range Variable Wall® Piping Calibration Blocks for Longseam Inspection

This set of two blocks has been specially designed for situations where ultrasonic examination tak
...
es place circumferentially or “around the curve” for applications such as inspection of longseam welds in piping and vessels. Jointly designed by PH Tool and Holloway NDT, the blocks feature side-drilled hole reflectors positioned within novel compound-curves, each with multiple thickness steps. The unique design results in a light and compact form factor, with the largest block weighing just over 15 lbs and spanning just 16 ¾”. Coverage is provided for nominal pipe sizes from 6” to 10” in the small block and 12” to a whopping 56” in the large block. The compound OD curves and multiple wall thicknesses actually exceed ASME requirements, providing exceptional signal response and measurement accuracy. Properly managing beam divergence off the ID and stabilizing beam skew due to probe rocking is essential to maintaining a consistent calibration when inspecting lengthwise on curved surfaces.
>

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window