EN 10160 interpretation

Dear Expert;
We work on UT test for plate.
For EN 10160 class S1;
standard say permisible defect size and number of defects are following :

Area
Defect mm² Max. number of defects
Within 1 m²
100 - 1000 15

My question is if there is 15 defects in this range and also there is some defect that's sizes is lower than 100mm². At this condition this plate is acceptable or not ?

Thank you



 
 Reply 
 
bobsudharmin
Engineering, Reliability and Integrity Eng
Shell Malaysia Trading, Malaysia, Joined Jan 2008, 54

bobsudharmin

Engineering, Reliability and Integrity Eng
Shell Malaysia Trading,
Malaysia,
Joined Jan 2008
54
05:15 Jul-28-2008
Re: EN 10160 interpretation
Isn't the statement clear enough? Defects of size less than 100mm2 is not to be considered. The criteria would be ACCEPT if it does not exceed 15 in nos.

best regards,
bobsudharmin


----------- Start Original Message -----------
: Dear Expert;
: We work on UT test for plate.
: For EN 10160 class S1;
: standard say permisible defect size and number of defects are following :
: Area
: Defect mm² Max. number of defects
: Within 1 m²
: 100 - 1000 15
: My question is if there is 15 defects in this range and also there is some defect that's sizes is lower than 100mm². At this condition this plate is acceptable or not ?
: Thank you
------------ End Original Message ------------




 
 Reply 
 
Y.Karavacioglu
Y.Karavacioglu
07:57 Aug-01-2008
Re: EN 10160 interpretation
Dear Bobsudharmin,
Think the other condition
if there is 15 defects in this range and also there is some defect that's sizes is higher than 1000mm². At this condition this plate is acceptable or not ?

Thank you

----------- Start Original Message -----------
: Isn't the statement clear enough? Defects of size less than 100mm2 is not to be considered. The criteria would be ACCEPT if it does not exceed 15 in nos.
:
: best regards,
: bobsudharmin
:
: : Dear Expert;
: : We work on UT test for plate.
: : For EN 10160 class S1;
: : standard say permisible defect size and number of defects are following :
: : Area
: : Defect mm² Max. number of defects
: : Within 1 m²
: : 100 - 1000 15
: : My question is if there is 15 defects in this range and also there is some defect that's sizes is lower than 100mm². At this condition this plate is acceptable or not?
: : Thank you
------------ End Original Message ------------




 
 Reply 
 
bob sudharmin
Engineering, Reliability and Integrity Eng
Shell Malaysia Trading, Malaysia, Joined Jan 2008, 54

bob sudharmin

Engineering, Reliability and Integrity Eng
Shell Malaysia Trading,
Malaysia,
Joined Jan 2008
54
06:14 Aug-01-2008
Re: EN 10160 interpretation
Acceptance and rejection criteria are based on agreed boundaries, a draw off line. Like I said earlier, defects of lower than 100mm2 is not to be considered. To answer your query on the size of more than 1000mm2, it is an automatic reject. The code allows not more than 15 in nos of size in-between 100 to 1000mm2 in an area of 1 Meter sq.
If you have a single size of more than 1000mm2 in the same area with the 15 in nos, or in another area of the same plate for that matter, you no longer have to even consider the 15, since that single defect is above the permissible range.
Trust this clarifies.

----------- Start Original Message -----------
: Dear Bobsudharmin,
: Think the other condition
: if there is 15 defects in this range and also there is some defect that's sizes is higher than 1000mm². At this condition this plate is acceptable or not ?
: Thank you
: : Isn't the statement clear enough? Defects of size less than 100mm2 is not to be considered. The criteria would be ACCEPT if it does not exceed 15 in nos.
: :
: : best regards,
: : bobsudharmin
: :
: : : Dear Expert;
: : : We work on UT test for plate.
: : : For EN 10160 class S1;
: : : standard say permisible defect size and number of defects are following :
: : : Area
: : : Defect mm² Max. number of defects
: : : Within 1 m²
: : : 100 - 1000 15
: : : My question is if there is 15 defects in this range and also there is some defect that's sizes is lower than 100mm². At this condition this plate is acceptable or not ?
: : : Thank you
------------ End Original Message ------------




 
 Reply 
 
Robert Emrys Sidney Long
Robert Long, Denmark, Joined Apr 2020, 1

Robert Emrys Sidney Long

Robert Long,
Denmark,
Joined Apr 2020
1
14:49 Apr-02-2020
Re: EN 10160 interpretation
In Reply to bob sudharmin at 06:14 Aug-01-2008 .

en 10160
size of the discontinuity.
for S2 is it 5 or 8 mm max.

with a double transducer

it is a little ambiguous.
states 5 mm and the all less than 8 should be disregarded.
thankyou

 
 Reply 
 
THEO MICOTTIS
Italy, Joined May 2013, 82

THEO MICOTTIS

Italy,
Joined May 2013
82
17:15 Apr-02-2020
Re: EN 10160 interpretation
In Reply to Robert Emrys Sidney Long at 14:49 Apr-02-2020 .

Using double transducer probe the reference is always 5mm regardless of qualty class

 
 Reply 
 
Marcelo
Engineering,
Brazil, Joined Sep 2015, 7

Marcelo

Engineering,
Brazil,
Joined Sep 2015
7
13:18 Apr-06-2020
Re: EN 10160 interpretation
In Reply to Robert Emrys Sidney Long at 14:49 Apr-02-2020 .

With double cristal If you use a calibration FBH >5mm it may have full reflection of the echo due to the size of the FBH surface. This surface is to big and behaves like a bottom surface for this probe.

 
 Reply 
 

Product Spotlight

PAUT Probes

Typical Phased Array probes have frequencies between 1MHz and 20MHz and the number of wafers is 10
...
to 128. M2 Electronics offers customers conventionally ultrasound probes and the ability to provide high-precision Phased Array Ultrasound Probes of up to 256 wafers. We can also customize the probe for our customers to meet the specific application requirements of the user.
>

MUSE Mobile Ultrasonic Equipment

The MUSE, a portable ultrasonic imaging system, was developed for in-field inspections of light-weig
...
ht structures. The MUSE consists of a motor-driven manipulator, a water circulation system for the acoustic coupling and a portable ultrasonic flaw detector (USPC 3010). The MUSE provides images of internal defects (A-, B-,C- and D-scan).
>

Wireless TOFD scanner

Quick, accurate and highly reproducible welds testing. The System operates wirelessly and is compat
...
ible with any type of Windows based Laptop, Desktop or Tablet.
>

AIS229 - Multipurpose Real Time System

Latest standard & automatic real time system developed by Balteau. The AIS229 has been designed to
...
do series inspection in a wide variety of industry. Composed of a shielded cabinet, 5 axis manipulator, x-ray generator and tubehead from 160kV to 225kV, a fl at panel & much more, the AIS229 is most certainly one of the most multipurpose RTR system available on the market.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window