02:32 Aug-08-2008
x ray

sir,
i would like to know upto which thickness in steel we can take radiodraphs with 2 percentage sensitivity?
what is its limitation?
In other way we have a 0.5mm thin welding (by just fusion without filler wire) Can we assess this weld quality by Xray radiography?
pl give me details?



    
 
 
Nigel
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
07:03 Aug-08-2008
Re: x ray
From your second question it seems you have access to X-ray equipment. With X-ray and some exposure parameter tweaking you should always be able to achieve 2% sensitivity. The only limitation will be the maximum steel thickness your X-ray equipment will penetrate. For your 0.5mm thick (steel?) weld, low kV X-ray can give you excellent results. If you wish to penetrate much thicker steel sections then, assuming you only have access to isotopes, you will need to make test shots to detemine the capability of the set-up to achieve 2% sensitivity. Factors affecting sensitivity are film and subject contrast and definition, so you need to consider all the factors which control those parameters, such as kV, filtering, scatter control, film speed. Dont you have exposure charts for the equipment and film you are working with?

Try a few test shots with the equipment materials you have to hand - seeing is believing and altering parameters and seeing the results will give you a good idea what you will need to do for production.

Regards

Nigel

----------- Start Original Message -----------
: sir,
: i would like to know upto which thickness in steel we can take radiodraphs with 2 percentage sensitivity?
: what is its limitation?
: In other way we have a 0.5mm thin welding (by just fusion without filler wire) Can we assess this weld quality by Xray radiography?
: pl give me details?
------------ End Original Message ------------




    
 
 
Afshun Agha Amiri
Afshun Agha Amiri
01:38 Aug-09-2008
Re: x ray
----------- Start Original Message -----------
Dear Friend,
In industrial RT For proper recommendations about energy selection and relevant thickness limitation on steel materials,you can refer to some code/standards(such as BS/EN1435 or old version of ASMEsec.5-before 1998),ASNT RT handbook,ASM handbook(vol.17)and others reference books.
2%sensitivity means : as a RT quality criteria and be checked by IQI,is 2% of test material thickness which is minimum detectable defect size through thickness.
you may obtain 2% sensitivity by x-ray or gamma-ray.
You can refer to ASM handbook(vol.17).


    
 
 
gowrishankar
R & D, scientist
ISRO, India, Joined Nov 2008, 34

gowrishankar

R & D, scientist
ISRO,
India,
Joined Nov 2008
34
06:31 Aug-12-2008
Re: x ray
Thank you sirs,
regards
gowrishankar

----------- Start Original Message -----------
: Dear Friend,
: In industrial RT For proper recommendations about energy selection and relevant thickness limitation on steel materials,you can refer to some code/standards(such as BS/EN1435 or old version of ASMEsec.5-before 1998),ASNT RT handbook,ASM handbook(vol.17)and others reference books.
: 2%sensitivity means : as a RT quality criteria and be checked by IQI,is 2% of test material thickness which is minimum detectable defect size through thickness.
: you may obtain 2% sensitivity by x-ray or gamma-ray.
: You can refer to ASM handbook(vol.17).
------------ End Original Message ------------




    
 
 
tj
tj
03:20 Nov-14-2008
Re: x ray
----------- Start Original Message -----------
: Dear Friend,
: In industrial RT For proper recommendations about energy selection and relevant thickness limitation on steel materials,you can refer to some code/standards(such as BS/EN1435 or old version of ASMEsec.5-before 1998),ASNT RT handbook,ASM handbook(vol.17)and others reference books.
: 2%sensitivity means : as a RT quality criteria and be checked by IQI,is 2% of test material thickness which is minimum detectable defect size through thickness.
: you may obtain 2% sensitivity by x-ray or gamma-ray.
: You can refer to ASM handbook(vol.17).
------------ End Original Message ------------

This could be misleading. You should never rely on your IQI as a gauge of minimum defect size. Your acceptance criteria and the quality indicator are two seperate issues.


    
 
 

Product Spotlight

I&T Nardoni

NDT, RESEARCH AND DEVELOPMENT OF NEW METHODOLOGY, TRAINING CENTER I&T NARDON INSTITUTE was idiated
...
in 1988 by Giuseppe Nardoni, now president and owner of the company, based on his multi-annual and international experience in the field of NDT.
>

MUSE Mobile Ultrasonic Equipment

The MUSE, a portable ultrasonic imaging system, was developed for in-field inspections of light-weig
...
ht structures. The MUSE consists of a motor-driven manipulator, a water circulation system for the acoustic coupling and a portable ultrasonic flaw detector (USPC 3010). The MUSE provides images of internal defects (A-, B-,C- and D-scan).
>

NEW! The PragmaPro Instrument Platform

The PragmaPro is based on a modular cartridge technology and supports various NDT instrument modal
...
ities such as UT, PAUT, ECT and many more. This new platform is based on a machined, powder-coated aluminum frame for shock-proofness, best sealing qualities and maximum heat dissipation. This is practical to extend the outdoor temperature range and/or to extend the power injected in the transducers. The PragmaPro is aiming at a very wide range of applications, such as weld scanning, corrosion mapping and composite testing.
>

AMIGO2

TSC Amigo2 - ACFM technology has developed a solid reputation for accurately detecting and sizing
...
surface-breaking cracks through paint and coatings. As the industry demands increased performance in speed, signal quality, and portability, it’s time for an evolution. It’s time for Amigo2.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window