x ray

sir,
i would like to know upto which thickness in steel we can take radiodraphs with 2 percentage sensitivity?
what is its limitation?
In other way we have a 0.5mm thin welding (by just fusion without filler wire) Can we assess this weld quality by Xray radiography?
pl give me details?



 
 Reply 
 
Nigel
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
07:03 Aug-08-2008
Re: x ray
From your second question it seems you have access to X-ray equipment. With X-ray and some exposure parameter tweaking you should always be able to achieve 2% sensitivity. The only limitation will be the maximum steel thickness your X-ray equipment will penetrate. For your 0.5mm thick (steel?) weld, low kV X-ray can give you excellent results. If you wish to penetrate much thicker steel sections then, assuming you only have access to isotopes, you will need to make test shots to detemine the capability of the set-up to achieve 2% sensitivity. Factors affecting sensitivity are film and subject contrast and definition, so you need to consider all the factors which control those parameters, such as kV, filtering, scatter control, film speed. Dont you have exposure charts for the equipment and film you are working with?

Try a few test shots with the equipment materials you have to hand - seeing is believing and altering parameters and seeing the results will give you a good idea what you will need to do for production.

Regards

Nigel

----------- Start Original Message -----------
: sir,
: i would like to know upto which thickness in steel we can take radiodraphs with 2 percentage sensitivity?
: what is its limitation?
: In other way we have a 0.5mm thin welding (by just fusion without filler wire) Can we assess this weld quality by Xray radiography?
: pl give me details?
------------ End Original Message ------------




 
 Reply 
 
Afshun Agha Amiri
Afshun Agha Amiri
01:38 Aug-09-2008
Re: x ray
----------- Start Original Message -----------
Dear Friend,
In industrial RT For proper recommendations about energy selection and relevant thickness limitation on steel materials,you can refer to some code/standards(such as BS/EN1435 or old version of ASMEsec.5-before 1998),ASNT RT handbook,ASM handbook(vol.17)and others reference books.
2%sensitivity means : as a RT quality criteria and be checked by IQI,is 2% of test material thickness which is minimum detectable defect size through thickness.
you may obtain 2% sensitivity by x-ray or gamma-ray.
You can refer to ASM handbook(vol.17).


 
 Reply 
 
gowrishankar
R & D, scientist
ISRO, India, Joined Nov 2008, 34

gowrishankar

R & D, scientist
ISRO,
India,
Joined Nov 2008
34
06:31 Aug-12-2008
Re: x ray
Thank you sirs,
regards
gowrishankar

----------- Start Original Message -----------
: Dear Friend,
: In industrial RT For proper recommendations about energy selection and relevant thickness limitation on steel materials,you can refer to some code/standards(such as BS/EN1435 or old version of ASMEsec.5-before 1998),ASNT RT handbook,ASM handbook(vol.17)and others reference books.
: 2%sensitivity means : as a RT quality criteria and be checked by IQI,is 2% of test material thickness which is minimum detectable defect size through thickness.
: you may obtain 2% sensitivity by x-ray or gamma-ray.
: You can refer to ASM handbook(vol.17).
------------ End Original Message ------------




 
 Reply 
 
tj
tj
03:20 Nov-14-2008
Re: x ray
----------- Start Original Message -----------
: Dear Friend,
: In industrial RT For proper recommendations about energy selection and relevant thickness limitation on steel materials,you can refer to some code/standards(such as BS/EN1435 or old version of ASMEsec.5-before 1998),ASNT RT handbook,ASM handbook(vol.17)and others reference books.
: 2%sensitivity means : as a RT quality criteria and be checked by IQI,is 2% of test material thickness which is minimum detectable defect size through thickness.
: you may obtain 2% sensitivity by x-ray or gamma-ray.
: You can refer to ASM handbook(vol.17).
------------ End Original Message ------------

This could be misleading. You should never rely on your IQI as a gauge of minimum defect size. Your acceptance criteria and the quality indicator are two seperate issues.


 
 Reply 
 

Product Spotlight

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

Surf-X® Array Probe

Introducing the Surf-X family of flexible Eddy Current Array (ECA) probes. Featuring unique multiple
...
coil sets and proprietary X-PROBE technology, Surf-X array probes can quickly and accurately handle a range of inspection applications, from inspecting corrosion or cracking in pipes, pressure vessels, or tanks, to assessing and sizing cracks in raised welds and friction stir welds.
>

NEOS III

NEOS III is Logos Imagings lightest DR system. With a built-in battery and internal wireless commu
...
nication, the NEOS III is perfect for users that want to quickly assess an item.
>

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window