coercive force

is the coercive force depend on the shape ? Provided two samples are made of the same material but different shape, you measure their coercive force could you end up with different results ?



 
 Reply 
 
Michel Couture
NDT Inspector,
consultant, Canada, Joined Sep 2006, 894

Michel Couture

NDT Inspector,
consultant,
Canada,
Joined Sep 2006
894
07:24 Sep-05-2008
Re: coercive force
In one word, YES! The lower the L/D Ratio (Length to Diameter), greater is the Coercive Force required to demagnetise the item. But shape is not the only factor. You also have to take into account such properties as heat treatment and alloying.

Cheerio's
Michel

----------- Start Original Message -----------
: is the coercive force depend on the shape ? Provided two samples are made of the same material but different shape, you measure their coercive force could you end up with different results ?
------------ End Original Message ------------




 
 Reply 
 
Catalin Mandache
Catalin Mandache
06:03 Sep-08-2008
Re: coercive force
The coercive force depends on so many micro structural factors that is very unlikely to have the same values for two identically shaped and alloyed samples.

----------- Start Original Message -----------
: is the coercive force depend on the shape ? Provided two samples are made of the same material but different shape, you measure their coercive force could you end up with different results ?
------------ End Original Message ------------




 
 Reply 
 
David Mackintosh
Engineering,
Acuren Group Inc., Canada, Joined Feb 2011, 85

David Mackintosh

Engineering,
Acuren Group Inc.,
Canada,
Joined Feb 2011
85
09:36 Sep-23-2008
Re: coercive force
----------- Start Original Message -----------
: is the coercive force depend on the shape ? Provided two samples are made of the same material but different shape, you measure their coercive force could you end up with different results ?
------------ End Original Message ------------

The coercive force, Hc, is the applied field required to demagnetize a material after saturation. It is a material property. (See for example http://www.ndt-ed.org/EducationResources/CommunityCollege/MagParticle/Physics/HysteresisLoop.htm and http://en.wikipedia.org/wiki/Coercive_force.) The magnetic H-field required to demagnetize a specific object depends on the material properties, the shape of the object, and the method of applying the field, among other things. Hc for the material and the demagnetizing force required for a specific object are related, but it is not always easy to calculate one from the other.

David



 
 Reply 
 

Product Spotlight

Ultrasonic Probe Recharacterization Service

NDT Systems offers a comprehensive Aftercare and Recharacterization Service for all our ultrasonic
...
probes. The Recharacterization Service is fully compliant with International ASTM E1065 Standard Guide (and other applicable standards) and offers complete documentation, traceable to the ASTM E1065 Standard. For more details or to schedule Recharacterization Services contact ndtsales@ndtsystems.com
>

ISONIC 3510

Very Powerful Superior Performance Extremely Portable Smart Phased Array Ultrasonic Flaw Detector an
...
d Recorder with 2 Conventional UT and TOFD Channels uniquely combines PA, single- and multi-channel conventional UT, and TOFD modalities providing 100% raw data recording and imaging. Suitable for all kinds of every-day ultrasonic inspections
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

Magnetic X-Ray Pipeline Crawler 6''-60''

Suitable to work in 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometer
...
s and provide the efficient inspection of the pipeline.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window