04:56 May-09-1998
straight beam probe
: Dear sir,

I am a master student in mechanical eng.My thesis subject is to produce
a straight beam UT probe.

I have produce a few prototypes of 1MHz and 2MHz probes.They will be direct contact probes
I have used tungsten-epoxy mixtures( 50% Tungsten-50% Epoxy in volume), plastic for housing.
But I could not decide on the protecting layer.Then I have tried steel and brass plates ( 2 lambda thick)
but I got very high internal noise(Grass)and I could not eliminate it.I am not sure about the
source of these noise. So could you answer the below questions?
1- Are the source of these noises electrical or mechanical(assembly)?
2- What can I do to eliminate these noises?
3- Should I use screening? If so, how?
4- The electrical impedance is: For 1MHz R=6.1 ohm
X=-49 ohm
For 2MHZ R=76 ohm
X=13 ohm
Do I need to impedance matching for above values to 50 ohm generator?



 
 Reply 
 
gazi yuksel
gazi yuksel
05:01 May-09-1998
straight beam probe
: Dear sir,

I am a master student in mechanical eng.My thesis subject is to produce
a straight beam UT probe.

I have produce a few prototypes of 1MHz and 2MHz probes.They will be direct contact probes
I have used tungsten-epoxy mixtures( 50% Tungsten-50% Epoxy in volume), plastic for housing.
But I could not decide on the protecting layer.Then I have tried steel and brass plates ( 2 lambda thick)
but I got very high internal noise(Grass)and I could not eliminate it.I am not sure about the
source of these noise. So could you answer the below questions?
1- Are the source of these noises electrical or mechanical(assembly)?
2- What can I do to eliminate these noises?
3- Should I use screening? If so, how?
4- The electrical impedance is:
For 1MHz R=6.1 ohm X=-49 ohm
For 2MHZ R=76 ohm X=13 ohm
Do I need to impedance matching for above values to 50 ohm generator?

Thanks


 
 Reply 
 
Linas Svilainis
R & D,
Kaunas University of Technology, Lithuania, Joined Nov 1998, 67

Linas Svilainis

R & D,
Kaunas University of Technology,
Lithuania,
Joined Nov 1998
67
01:31 May-11-1998
Re: straight beam probe

: I am a master student in mechanical eng....
Good to hear some news about ultrasonic activities in
Turkey. Are you employed by ASELSAN?

: I have produce a few prototypes of 1MHz and 2MHz probes.They will be direct contact probes
Could you specify more of your design?
It's just a simple piston or something else?
Is it sransmitting or receiving? Both?

: I have used tungsten-epoxy mixtures( 50% Tungsten-50% Epoxy in volume),
But it's very hard to mix at such proportions!

:plastic for housing.
Nevertherless, you should use shielding inside
to protect your active electrode from noise induction

: But I could not decide on the protecting layer.Then I have tried steel and brass plates ( 2 lambda thick)
We think this will cause high resonant response here.
Once heavily backing is applied, by sacrifying the
efficiency of transducer bandwidth will be improved.
So, the 2 lambda is no longer valid and you might have,
say, 10 lambda...So,you have to be aware of resonance
effects...

: but I got very high internal noise(Grass)and I could not eliminate it.I am not sure about the
: source of these noise.
How much of "grass" have you got?

: So could you answer the below questions?
: 1- Are the source of these noises electrical or mechanical(assembly)?
In case the noise is decreasing with distance, then it
is of mechanicall matter. Is this noise present when you excite
the transducer or when it's just receiving?
If it does not depend on distance/excitation
is fluctuating and it's 1MHz transducer, you're
probably picking up SW station signals.

: 2- What can I do to eliminate these noises?
Check our the noise source, whether it's due to
design or electrical. Check out your PRF.
Try shielding.

: 3- Should I use screening? If so, how?
attach additional copper cylinder around the
backing and electrodes (aware of electrical contact!).

: 4- The electrical impedance is:
: For 1MHz R=6.1 ohm X=-49 ohm
: For 2MHZ R=76 ohm X=13 ohm
: Do I need to impedance matching for above values to 50 ohm generator?
Try to change the connecting cable length.
Does anything has changed? If you have sharp
exciting spike, the reflections traveling inside the
cable can produce such effect. But since your generator
is 50ohm, it should not be expressed so much.

Good luck



 
 Reply 
 

Product Spotlight

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

iProbe - USB Phased Array Probe

Turn your PC, Laptop or Tablet into a powerful 32:64 Phased Array system with our USB powered Phas
...
ed Array iProbe and inspection software package. Perform affordable weld inspection and corrosion mapping inspections with this powerful but small system. The probe has a 100V pulser, 8K PRF, dual axis encoder input and achieves outstanding signal quality.
>

ISAFE3 Intrinsically Safe Sensor System

ISAFE3 intrinsically safe sensor system of Vallen Systeme is especially targeted at the petrochemica
...
l - as well as oil and gas transportation industry. The sensor system is designed for permanent monitoring or periodic inspection tasks. Sensors are available for different AE-frequency ranges optimized for corrosion and fatigue crack detection and other applications. The ISAFE 3 sensor system consists of an AE-sensor (model ISAS3) certified according to ATEX/IEC for installation in zone 0, gas group IIC, IP68, 20 to +60 °C, and a signal isolator (model SISO3) certified for installation in zone 2. An ISAS3 sensor can be mounted in atmosphere or submerged, e.g. in water or crude oil. It is supported by mounting tools for temporary (magnets) or permanent (welded) installation. ISAFE3 supports automatic sensor coupling test and can be used with any AE signal processor supporting 28V supply at 90 mA peak, e.g. Vallen Systeme ASIP-2/A.
>

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window