Elect.Impedance of PZT Pickup


Dear colleagues

For economical purpose I had construct a simple PZT pickup
From PZT -4 Material of Fundamental frequency 2MHZ
When I measured the Electrical Impedance of Pickup against frequency
(Over frequency range 10 KHz to 4 MHZ )
I was found three Peaks
1- first at 250 kHz
2- second at 850 kHz
3- third at 3300 kHz
I Haven't explanation for this result, may be any colleague
has an explanation
Thank you very much



    
 
 
Jim Barshinger
Engineering
General Electric Global Research Center, USA, Joined Aug 2000, 8

Jim Barshinger

Engineering
General Electric Global Research Center,
USA,
Joined Aug 2000
8
05:44 Sep-18-2000
Re: Elect.Impedance of PZT Pickup
:
: Dear colleagues

: For economical purpose I had construct a simple PZT pickup
: From PZT -4 Material of Fundamental frequency 2MHZ
: When I measured the Electrical Impedance of Pickup against frequency
: (Over frequency range 10 KHz to 4 MHZ )
: I was found three Peaks
: 1- first at 250 kHz
: 2- second at 850 kHz
: 3- third at 3300 kHz
: I Haven't explanation for this result, may be any colleague
: has an explanation
: Thank you very much

The first and second peaks are most likely lateral resonances in the material. If you only have one lateral dimension, for instance a circular or square ceramic, I would guess that the 250 kHz peak is the fundamental lateral resonance and the 850 kHz peak is the third harmonic of the lateral resonance. The 3300 kHz peak should be your fundamental thickness mode resonance. I'm guessing that you are measuring the parallel resonance which will be a point of maximum real impedance. The series resonance should occur slightly lower in frequency and is a point of minimum real impedance. This should occur at 2 MHz (if you have ground your ceramic to the correct thickness). The center frequency that the device will run at depends on whether your recieving electronics matches better with the series or parallel resonance.

Good luck,
Jim





    
 
 

Product Spotlight

GUL QSR1® Scanning

How do you measure pipe wall thickness without direct access to the area? QSR® Scanning - Guide Wav
...
e Quantitative Short Range Scanning.
>

NEW Wheel Type Phased Array Probe

DOPPLER NEW Wheel Type Phased Array Probe, more stable, new tyre makes lesser acoustic attenuation
...
, much lighter makes easier to handle, more slim size, magnetic and mechanical encoder optional etc...more
>

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Ultrasonic tomograph for imaging of concrete structures А1040 MIRA

Applicable for concrete inspection allowing imaging of the internal structure of objects from conc
...
rete, reinforced concrete, different stones. The operation applies pulse-echo technique at one-side access to the object. The instrument is feasible for concrete inspection for searching conduct ducts, conduits, detection of foreign inclusions, holes, honeycombing, cracks and other concrete defects.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window