03:00 Oct-24-2000
Doubt

Dear experts:
I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
does it cause some anormality to manipulate it?

Kind regards,

Mario Belo


    
 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 176

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
176
09:25 Oct-25-2000
Re: Doubt
reply by Udo Schlengermann

Hello,

We are proud to see an instrument which is more than 20 years out of production in operation:
The USM2 had a calibrated decibel-attenuator (gain) stepped in 2-dB-steps.
To set an echo from a reference reflector exactly to a reference line an additional uncalibrated gain is necessary, if the 2-dB-steps do not fit exactly.
This uncalibrated gain is changed by the screw inside the dB-knob.
It's right position gives the maximum uncalibrated gain, it's left position the minimum uncalibrated gain.
Usually this uncalibrated gain should be set to its max-position.

Kind regards

Udo Schlengermann
+++++++

: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo




    
 
 
Bill Waldron
Sales
GE Inspection Technologies, USA, Joined Feb 2001, 2

Bill Waldron

Sales
GE Inspection Technologies,
USA,
Joined Feb 2001
2
02:48 Oct-25-2000
Re: Doubt
: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo

Mario,
This screw gives 10 dB of gain adjustment and does not affect linearity. Its intended use is to provide a means to fine adjust an echo to a reference level.
(As the USM2 attenuator is in 2 dB steps, it may not be possible to get an echo to an exact level. In this case you would use the small screw adjustment to fine tune the gain level.
Bill Waldron


    
 
 

Product Spotlight

FlexoFORM™

The FlexoFORM™ solution integrates flexible phased array probe technology in a scanner to solv
...
e the challenges that come with inspecting pipe elbows. When combined with a water column, this solution can be used to collect easy-to-interpret data on elbows with diameters ranging from 4.5 in. OD up to flat and offers many benefits.
>

ISONIC 3510

Very Powerful Superior Performance Extremely Portable Smart Phased Array Ultrasonic Flaw Detector an
...
d Recorder with 2 Conventional UT and TOFD Channels uniquely combines PA, single- and multi-channel conventional UT, and TOFD modalities providing 100% raw data recording and imaging. Suitable for all kinds of every-day ultrasonic inspections
>

Lyft™: Pulsed Eddy Current Reinvented

PEC Reinvented—CUI Programs Redefined Corrosion under insulation (CUI) is possibly the greatest u
...
nresolved asset integrity problem in the industry. Current methods for measuring wall thickness with liftoff, without removing insulation, all have severe limitations. Eddyfi introduces Lyft — a reinvented, high-performance pulsed eddy current (PEC) solution. The patent- pending system features a state-of-the-art portable instrument, real- time C-scan imaging, fast data acquisition with grid-mapping and dynamic scanning modes, and flexibility with long cables. It can also scan through thick metal and insulation, as well as aluminum, stainless steel, and galvanized steel weather jackets. Who else but Eddyfi to reinvent an eddy current technique and redefine CUI programs. Got Lyft?
>

I&T Nardoni

NDT, RESEARCH AND DEVELOPMENT OF NEW METHODOLOGY, TRAINING CENTER I&T NARDON INSTITUTE was idiated
...
in 1988 by Giuseppe Nardoni, now president and owner of the company, based on his multi-annual and international experience in the field of NDT.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window