03:00 Oct-24-2000
Doubt

Dear experts:
I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
does it cause some anormality to manipulate it?

Kind regards,

Mario Belo


 
 Reply 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 183

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
183
09:25 Oct-25-2000
Re: Doubt
reply by Udo Schlengermann

Hello,

We are proud to see an instrument which is more than 20 years out of production in operation:
The USM2 had a calibrated decibel-attenuator (gain) stepped in 2-dB-steps.
To set an echo from a reference reflector exactly to a reference line an additional uncalibrated gain is necessary, if the 2-dB-steps do not fit exactly.
This uncalibrated gain is changed by the screw inside the dB-knob.
It's right position gives the maximum uncalibrated gain, it's left position the minimum uncalibrated gain.
Usually this uncalibrated gain should be set to its max-position.

Kind regards

Udo Schlengermann
+++++++

: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo




 
 Reply 
 
Bill Waldron
Sales
GE Inspection Technologies, USA, Joined Feb 2001, 2

Bill Waldron

Sales
GE Inspection Technologies,
USA,
Joined Feb 2001
2
02:48 Oct-25-2000
Re: Doubt
: Dear experts:
: I would like to know, in what cases I should use the small screw located inside the control of gain of my USM 2?
: does it cause some anormality to manipulate it?
:
: Kind regards,
:
: Mario Belo

Mario,
This screw gives 10 dB of gain adjustment and does not affect linearity. Its intended use is to provide a means to fine adjust an echo to a reference level.
(As the USM2 attenuator is in 2 dB steps, it may not be possible to get an echo to an exact level. In this case you would use the small screw adjustment to fine tune the gain level.
Bill Waldron


 
 Reply 
 

Product Spotlight

CIVA 2020 UT Module

CIVA NDE Simulation Software is the world leader of NDT Simulation. The UT simulation Module incl
...
udes: - "Beam computation": Beam propagation simulation - "Inspection Simulation": Beam interaction with flaws or specimens The user can simulate a whole inspection process (pulse echo, tandem or TOFD) with a wide range of probes (conventional, Phased- arrays or EMAT), components, and flaws.
>

Magnetic X-Ray Pipeline Crawler 6''-60''

Suitable to work in 6''-60''diameter with speed of up to 18m/min, Max. moving diatance 5 kilometer
...
s and provide the efficient inspection of the pipeline.
>

ISONIC 3510

Very Powerful Superior Performance Extremely Portable Smart Phased Array Ultrasonic Flaw Detector an
...
d Recorder with 2 Conventional UT and TOFD Channels uniquely combines PA, single- and multi-channel conventional UT, and TOFD modalities providing 100% raw data recording and imaging. Suitable for all kinds of every-day ultrasonic inspections
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window