00:16 Nov-02-2000
Corrections calculation

Dear experts:

Is it possible, to determine the transference corrections (surface and attenuation) from a reference block with endless wall (V1/5) to an unknown flaw (smaller than the beam) in a certain piece with different acoustic characteristics and a different surface, which does not present backwall, through the correspondent AVG (DGS) diagram?
An example would be welcome.

Best regards,

Mario Belo



 
 Reply 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 179

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
179
09:35 Nov-14-2000
Re: Corrections calculation
reply by Udo Schlengermann

If you calibrate on a block which has different material and surface than the test object, this is not possible, for the following reasons:
Four different losses determine the signal amplitude:
1. divergence - depending on the frequency, the roughness and the size of the contact area.
2. attenuation - in reference block and test object.
3. transmission losses - if the materials are different for both objects (acoustical impedances).
4. reflectivity of reflectors - calibration reflector and detected discontinuity.

The DGS diagram is designed for a full size contact (effective transducer diameter) and a material without attenuation.
Most standards therefore allow DGS method only if the curvature compared to transducer size is below a limit.
Transfer losses on curved, rough objects therefore are only valid for a given contour and roughness, and a defined distance of reflectors, i.e. they change with distance, if attenuation is not comparably equal on both objects.
Before using transfer losses, they have to be determined on a test block having the same curvature and the same roughness as the test object, containing the same reflector (e.g. backwall) at the same distance than your calibration block. Then losses no 4. and 3. are constant in both cases and the remaining difference to the DGS-curve is the transfer loss (attenutation and transmission losses).
It is therefore necessary to have at least one reflector of known size and distance in the test object. You cannot determine transfer losses using an unknown reflector.

Kind regards

Udo Schlengermann
++++++
: Dear experts:

: Is it possible, to determine the transference corrections (surface and attenuation) from a reference block with endless wall (V1/5) to an unknown flaw (smaller than the beam) in a certain piece with different acoustic characteristics and a different surface, which does not present backwall, through the correspondent AVG (DGS) diagram?
: An example would be welcome.

: Best regards,

: Mario Belo




 
 Reply 
 

Product Spotlight

iProbe - USB Phased Array Probe

Turn your PC, Laptop or Tablet into a powerful 32:64 Phased Array system with our USB powered Phas
...
ed Array iProbe and inspection software package. Perform affordable weld inspection and corrosion mapping inspections with this powerful but small system. The probe has a 100V pulser, 8K PRF, dual axis encoder input and achieves outstanding signal quality.
>

GEKKO - Portable Phased Array Testing with TFM in Real-Time

The portable phased array testing system GEKKO provides 64 parallel test channels. On creating testi
...
ng parameters the operator is assisted by the CIVA software. Due to its modular set-up the GEKKO instrument is suitable for operators of all skill levels.
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

Compact NDT inspection-heads for measurements with active thermography

The compact inspection head is suitable for thermographic ndt tasks. The uncooled infrared camera
...
is specially developed for NDI-tasks and offers a thermal sensitivity until now known only from thermal imagers with cooled detector. All required components and functions are integrated into the inspection-head. You will only need an ethernet cable to connect the sensor with the evaluation system.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window