Is the GE specification good and still relevant?

The GE spec describes the probe's and equipment's capability
for its near to surface resolution.
See also GE Spec in UT A-Z.

For my opinion this specification let to much parameters open,
for instant there is not any dB value mentioned about the
signal noise ratio at the flaw's echo. Is this true?

Are there better standards available?


    
 
 
Robert A. Dau
Engineering
Milky Way Jewels, USA, Joined Nov 1998, 40

Robert A. Dau

Engineering
Milky Way Jewels,
USA,
Joined Nov 1998
40
09:49 Sep-11-1996
Re: Is the GE specification good and still relevant?
Rolf -

Near surface resolution remains a difficult area to
specify because the actual parameters you need depend on
what size defect you need to detect and on the signal-to-
noise you need to reliably detect it. Certainly the GE
specification will not be adequate in some cases, it
will be in others. So my answer is it depends.

One very good way to make a specification for a
particular test is to pin everything down. Specify the
near surface resolution in terms of a reference FBH a
certain distance from the surface and the dB of signal-
to-noise you need. The GE specification is of this type
but not completely rigorous because it doesn't specify
how you measure SN. If you want to detect 0.4 mm FBH 1.2
mm into steel. The GE spec. is good although putting a
more "scientific" definition on the two wave amplitudes
is desirable.

If you want to apply the GE spec. to another material
that is more attenuative for example, then the GE spec.
will be inappropriate. Specifying the near surface
independently from the application is as far as I can
see beyond our current knowledge. The "do everything in
steel and translate to other materials method" is of
limited use to most ultrasonic users because there are
no "translation" codes available to do that. Indeed
there is not much data on attenuation as a function of
frequency out "there."

I might also add that near surface resolution is more
complex than most ultrasonic measurements because it is
a function of almost every parameter you can think off.
Center frequency, band width (damping), material
attenuation, element diameter, focal length, instrument
setting, and water path all change near surface
resolution. So I think you are stuck with using the GE
Spec. and "translating" it experimentally to your needs
or of creating a spec. specific to your own needs.

Regards,
Rocky

: The GE spec describes the probe's and equipment's capability
:for its near to surface resolution.
: See also GE Spec in UT A-Z.

: For my opinion this specification let to much parameters open,
: for instant there is not any dB value mentioned about the
: signal noise ratio at the flaw's echo. Is this true?

: Are there better standards available?




    
 
 

Product Spotlight

iProbe - USB Phased Array Probe

Turn your PC, Laptop or Tablet into a powerful 32:64 Phased Array system with our USB powered Phas
...
ed Array iProbe and inspection software package. Perform affordable weld inspection and corrosion mapping inspections with this powerful but small system. The probe has a 100V pulser, 8K PRF, dual axis encoder input and achieves outstanding signal quality.
>

High-performance Linear Phased Array Probes

Available to order from stock in a range of 5MHz – 7.5MHz and from 16 to 64 elements. Designed w
...
ith piezo-composite elements, Phoenix phased array probes provide high-resolution imaging to maximise sensitivity; accurate ultrasonic detection and sizing of defects in welds; and effective corrosion mapping. Housed in a rugged stainless steel case for on-site industrial NDT applications.
>

NEW! The PragmaPro Instrument Platform

The PragmaPro is based on a modular cartridge technology and supports various NDT instrument modal
...
ities such as UT, PAUT, ECT and many more. This new platform is based on a machined, powder-coated aluminum frame for shock-proofness, best sealing qualities and maximum heat dissipation. This is practical to extend the outdoor temperature range and/or to extend the power injected in the transducers. The PragmaPro is aiming at a very wide range of applications, such as weld scanning, corrosion mapping and composite testing.
>

FlexoFORM™

The FlexoFORM™ solution integrates flexible phased array probe technology in a scanner to solv
...
e the challenges that come with inspecting pipe elbows. When combined with a water column, this solution can be used to collect easy-to-interpret data on elbows with diameters ranging from 4.5 in. OD up to flat and offers many benefits.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window