Dear Experts,

I have some queries regarding the ultrasonic resolution. i did some study on ultrasonic transducer. i placed a transmitter & at the other end a receiver probe on the verticle column surface (non-invasive, with couplant) consist of water. my reason is to study/measure the air bubbles velocity released from the bottom of the column.

if my freq is 40kHz, and taking v= 1500m/s (water), from v = f*lamda, the wavelength of the sound is 37.5mm.

so, what is the minimum bubble size could be detected by the receiver probe? is it equivalent to the wavelength? because i did try smaller than the wavelength (25mm) and the receiver probe still gives the response.

thank you

    
 
 
Wieslaw Bicz
Engineering,
PBP Optel sp. z o.o., Poland, Joined Feb 2009, 247

Wieslaw Bicz

Engineering,
PBP Optel sp. z o.o.,
Poland,
Joined Feb 2009
247
19:12 Sep-15-2009
Re: spatial resolution
In Reply to Hafiz at 14:31 Aug-22-2009 (Opening).

The detection of single bubbles has nothing to do with resolution. This would be the case, if you would be interested in differentiation between two bubbles situated close to each other. In you case the question will be: how strong is the signal reflected by the bubble and what is the sensitivity of your probe and electronics. The wave length is not significant in this case. It can be much larger than the bubble size.

    
 
 

Product Spotlight

PROlineTOP Plug & Play Ultrasonic inspection device

As Plug & Play solution it units all control and operation elements in a small housing and therefo
...
re replaces the typical control cabinet...
>

Echomac® PA Rail Wheel Tester

MAC introduces two versions of Rail Wheel Testers using ECHOMAC® PA platform based on customer requ
...
irements.
>

Research and Applications Development For NDT

The Research and Applications Development (RAD) group is a newly formed team within Acuren dedicat
...
ed to tackling challenging inspection problems. Our focus is the development of novel, field deployable, advanced inspection techniques for use in cases where standard NDT methods are ineffective. We don't wait for new innovations, we engineer them. From concept to commissioning.
>

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window