20:16 Dec-15-2009
DAC/AWS/TCG/DGS

Im so used to using a DAC I have forgotten everything mentioned about these other methods on my courses if they where even brought up.

I would like to find out more and if its possible to use some of them in certain cases to improve my inspection methods.

What is each best for?

Type of calibration block for each?

Any literature on the differences?

Thanks in advance.

 
 Reply 
 
Nigel Armstrong
Engineering, - Specialist services
United Kingdom, Joined Oct 2000, 1096

Nigel Armstrong

Engineering, - Specialist services
United Kingdom,
Joined Oct 2000
1096
08:49 Dec-19-2009
Re: DAC/AWS/TCG/DGS
In Reply to anjafo at 20:16 Dec-15-2009 (Opening).

Hi Tony

TCG is a time-corrected DAC so that equal dimension reflectors give equal amplitude responses for all sound path distances. Used for PAUT Sectorial scans where it would be otherwise impossible to set every angle and sound path to the same sensitivity level using DAC's.

ASTM E-1316: DGS (distance gain size-German AVG)—distance amplitude curves permitting prediction of reflector size compared to the response from a back surface reflection.

The probe manufacturer supplies data sheet diagams for each probe which shows the amplitude response curves from the backwall and a range of diameters of flat-bottom holes along the length of the soundfield.

Have a look at EN 583-2:2001 Sensitivity and range setting for excellent authoritative descriptions of DAC/TCG and DGS. You'll have to look at AWS D1.1. for instance for knowledge of their sensitivity setting requirements.

Knowledge of these techniques is desirable but will such knowledge really improve your inspection method? You use DAC because the Codes and standards you work to require you to assess indications to those DAC's. A report that a reflector was 3,5 mm equivalent FBH size to DGS would most probably be rejected.

I always liked "Adjust gain until 2mm grass apears acrosss the timebase." and let the technician do the rest in terms of indication characterisation. In my opinion techs raised on this method interrogated the indication envelope (that phrase is almost an anachronism now) the most and built up the best picture of an indication's size, shape and reflecting surfaces. Note even CIT's such as PA and TOFD still often require manual UT characterisation.

 
 Reply 
 
Roger Duwe
NDT Inspector, API-510, 570, 653
MISTRAS, USA, Joined Jan 2009, 148

Roger Duwe

NDT Inspector, API-510, 570, 653
MISTRAS,
USA,
Joined Jan 2009
148
03:35 Dec-21-2009
Re: DAC/AWS/TCG/DGS
In Reply to Nigel Armstrong at 08:49 Dec-19-2009 .

Ah Nigel, no nostalga for this boy. After I convert a nice, clean DAC curve into a TCG flat-line, life gets very simplified.
"Does the indication break the [flat, straight] reference line?"
if answer is no:
"Does the indication break the nice, straight -12/14dB line?"

I can perform those evaluations 12 hours into a tough day, whiling hanging in a harness, reaching to the back side of a big pipe. Today, most of the tough work is performed during the calibrations. Makes the field measurements & evaluations so very much easier.

 
 Reply 
 
Joe Buckley
Consultant, ASNT L-III, Honorary Secretary of BINDT
Level X NDT, BINDT, United Kingdom, Joined Oct 1999, 526

Joe Buckley

Consultant, ASNT L-III, Honorary Secretary of BINDT
Level X NDT, BINDT,
United Kingdom,
Joined Oct 1999
526
15:15 Dec-21-2009
Re: DAC/AWS/TCG/DGS
In Reply to Roger Duwe at 03:35 Dec-21-2009 .

The other applications where TCG has major advantages:

1. automated system - just detect a gate being broken.

2. Highly attenuative materials such as composites or cast iron, where the signal amplitude variation in the region of interest exceeds the 15 dB or so it is practical to evaluate on a constant gain A-scan.

From a code point of view TCG and DAC are effectively equivalent, so you can generally make things a lot easier without a downside. Most codes date back from the days before TCG was such a low cost option, so its rare to find it actually specified.

I have seen instruments that implement a Time controlled gain version of DGS, but I'm not sure about that one.

 
 Reply 
 
Michael Moles †2014 *1948
, Joined ,
15:22 Dec-21-2009
Re: DAC/AWS/TCG/DGS
In Reply to Roger Duwe at 03:35 Dec-21-2009 .

This is exactly why the ASME Code requests a TCG, not a DAC. With a TCG, you can just set your thresold at one pre-defined colour, and everything above it is rejectable; if you want to go -12/-14 dB down from there, you can set your palette accordingly. That way, one can tell at a glance if there is a rejectable defect or not, or a defect that requires further investigation.

Naturally, it is possible to make a DAC (or multiple DACs) for AUT and phased arrays, but there is a LOT more work involved, and a much higher chance of error.

Michael Moles

 
 Reply 
 
adex
adex
17:15 Dec-22-2009
Re: DAC/AWS/TCG/DGS
In Reply to Michael Moles †2014 *1948 at 15:22 Dec-21-2009 .

Hi,

i'm a beginner UT Tech and have a very small experience in Phased array and other advance method. I need your explanation upon your post.

as your post above ;

"This is exactly why the ASME Code requests a TCG, not a DAC."

can you give me the ASME codes for my referrences.


"With a TCG, you can just set your thresold at one pre-defined colour, and everything above it is rejectable; if you want to go -12/-14 dB down from there, you can set your palette accordingly. That way, one can tell at a glance if there is a rejectable defect or not, or a defect that requires further investigation"

I understand but i have a question on +dB and -dB which will represent colour after we set the threshold, is it the +dB and -dB actually represent the amplitudes of the echo because of the setting on Full wave rectifier.

If the answer is NO, what is it?

If the answer is YES, some says that the electric sinusodial graph is not same as ultrasonic pulse wave, is it true ?

If i have wrong information, please correct me.

and lastly, is it true that any equipment that using the colour coding (thresold at one pre-defined colour )for their references need to use TCG???

thank you in advance.

 
 Reply 
 
Michael Moles †2014 *1948
, Joined ,
22:02 Dec-22-2009
Re: DAC/AWS/TCG/DGS
In Reply to adex at 17:15 Dec-22-2009 .

With pleasure. ASME currrently has five Code Cases in Section V Article 4, covering manual and encoded phased array scanning: 2541, 2557, 2558, 2599 and 2600. If you wait until the summer of 2010, we should have a Mandatory Appendix on phased arrays.

There are related practices in ASTM: E-2491 on setting up phased arrays, and E-2700, a procedure for contact weld inspections.

 
 Reply 
 
Krystel
Krystel
07:23 Mar-10-2015
Re: DAC/AWS/TCG/DGS
In Reply to Michael Moles †2014 *1948 at 15:22 Dec-21-2009 .

Michael Moles, what is the uniques advantage of your OmniScan MX2 over other Phased Array Technology?

My colleague is planning to buy phased array equipment.

Thank you.

 
 Reply 
 
Nicholas Bublitz
Other, Phased Array Analysist
VeriPhase, USA, Joined May 2010, 128

Nicholas Bublitz

Other, Phased Array Analysist
VeriPhase,
USA,
Joined May 2010
128
14:00 Mar-10-2015
Re: DAC/AWS/TCG/DGS
In Reply to Krystel at 07:23 Mar-10-2015 .

Krystel
I am with Olympus. My colleague and friend Michael has passed on, so I will answer your query on Olympus' behalf.
I recommend you get a demonstration of the multiple units out in the market and confirm your needs (environment, ease of use, features etc.). We could go on and on about perceived advantages. and another manufacturer will do the same on a forum like this. We do benefit from making the OmniScan platform and being in the phased array game for a long time. You can see the OmniScan page (and our other phased array instruments) here and also sign up for some one to contact you or find your representative on our web page.

http://www.olympus-ims.com/en/phasedarray/
http://www.olympus-ims.com/en/contact-us/

 
 Reply 
 

Product Spotlight

Immersion systems

ScanMaster ultrasonic immersion systems are designed for high throughput, multi shift operation in a
...
n industrial or lab environment. These fully integrated systems provide various scanning configurations and incorporate conventional and phased arrays technologies to support diverse applications, such as inspection of disks, bars, shafts, billets and plates. All of ScanMaster immersion systems are built from high accuracy scanning frames allowing for scanning of complex parts and include a multi-channel ultrasonic instrument with exceptional performance. The systems are approved by all major manufacturers for C-scan inspection of jet engine forged discs. Together with a comprehensive set of software modules these flexible series of systems provide the customer with the best price performance solutions.
>

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

EKOSCAN Phased Array

In order to always fit your needs, EKOSCAN can manufacture any type of UT transducer, either convent
...
ional or Phased Array. As an ISO 9001: 2015 certified company, EKOSCAN is extremely careful as far a material selection and manufacturing processes are concerned. Our probes guarantee our customers the benefits of latest innovations regarding piezo-composite, backing, impedance adaptation layer, etc. Specific probes for hostile environment: high temperature, high pressure, corrosive environment,etc. Specific probes designed to fit your specific application: optimization of every parameter to guarantee you the best detection.
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window