SAFT

Dear Sir

I would like to know if SAFT (Synthetic Aperture Focusing Technique) used in ultrasonics is the same as SAR (Synthetic Aperture Radar) and if SAFT requires coherent pulses to work.

If it does not use coherent pulses, where can I get more information on the equations used (since it cannot use diffraction equations).

If it does use coherent pulses, are there any manufacturers of industrical UT equipment which make coherent pulsers.

Best regards
Philippe Rubbers ( Philippe.Rubbers@eskom.co.za )


 
 Reply 
 
M. Schickert
R & D,
Institute for Materials Research and Testing, Germany, Joined Jan 2000, 18

M. Schickert

R & D,
Institute for Materials Research and Testing,
Germany,
Joined Jan 2000
18
00:05 Mar-18-2002
Re: SAFT
SAFT/SAR: SAFT has been developed in analogy to SAR (also adopting principles from optical holography and geophysics), so the algorithms are similiar except that in SAR usually far field approximations are made to accelerate computation. For synthetic focusing, all ultrasonic imaging techniques (SAFT, SAR, holography, tomography) require coherent superposition of received signals, although special incoherent methods exist.

Coherent Pulses: The term "coherent" applies to continuous waves in the first place. When using short pulses as a container for a broad frequency spectrum, the temporal relation of pulses determines the phase relation of the waves and therefore their coherence. So all one has to do in an imaging experiment is to assure fixed temoral relationship between the transmitted and each received signal. This can be done by triggering A/D-conversion by the transmitted pulse. For increased accuracy, the instruments clocks involved should be coupled. It should be noted that coherent imaging also requires coherent propagation of pulses in the medium imaged. Problems might occur in, e.g., fluctuating and strongly scattering media.


 
 Reply 
 

Product Spotlight

OmniScan™ X3 flaw detector

The OmniScan X3 flaw detector is a complete phased array toolbox. Powerful tools, like total focus
...
ing method (TFM) images and advanced visualization capabilities, enable you to complete your inspection with greater confidence.
>

ISONIC 3510

Very Powerful Superior Performance Extremely Portable Smart Phased Array Ultrasonic Flaw Detector an
...
d Recorder with 2 Conventional UT and TOFD Channels uniquely combines PA, single- and multi-channel conventional UT, and TOFD modalities providing 100% raw data recording and imaging. Suitable for all kinds of every-day ultrasonic inspections
>

Ultrasonic Squirter/Gantry System

TecScan’s NDT Ultrasonic Gantry Systems are industrial Squirter scanners designed for non-destru
...
ctive quality testing and raster scanning of large structures and parts. The Gantry Systems are usually composed of an Industrial Mechanical Scanner and a Control Room with System Workstation.
>

Combination of Digital Image Correlation and Thermographic Measurements

The combination of measuring results from the digital image correlation (ARAMIS, DIC) and temperat
...
ure measuring data from infrared cameras permits the simultaneous analysis of the thermal and mechanical behavior of test specimens in the materials and components testing field.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window