Controle Mesure Systemes
Contrôle Mesure Systèmes (CMS) is a leader in Non Destructive Testing (NDT) with a complete range of products for eddy current and ultrasonic i ...
822 views
Technical Discussions
isra
Consultant
isr, Brazil, Joined Apr 2002, 3

isra

Consultant
isr,
Brazil,
Joined Apr 2002
3
09:01 Apr-17-2002
Urgent!!!

the following data was obtained
1- the peak number inreases with the increase of transducer (probe ) diameter when the other parameter are constant.
2- The band width increases with decrease of transducer (probe) diameter when the other parameters are constant.

please help getting an interpritation for these two results.


 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1286

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1286
00:01 Apr-17-2002
Re: Urgent!!!
Some questions must be answered before this urgent matter can be resolved.
What do you mean by Peak Number.
Is it the number of peaks observed in the near zone?
Is it the peak frequency?
Is it the peak voltage?
Or do you mean centre frequency?
Or something else?

You then state that the bandwidth increases with diameter and claim that all other parameters remain constant.
I SUSPECT the effects of mechanical damping are the main cause of bandwidth variation. But unless you know that the entire construction process of the probe is identical it will be difficult to assert that ALL other parameters are constant.
Perhaps there are parameters varying that you have not considered.
Are these probes contact or immersion evaluated?

Ed

: the following data was obtained
: 1- the peak number inreases with the increase of transducer (probe ) diameter when the other parameter are constant.
: 2- The band width increases with decrease of transducer (probe) diameter when the other parameters are constant.
.
: please help getting an interpritation for these two results.
.



 
 Reply 
 
ISRA
Consultant
isr, Brazil, Joined Apr 2002, 3

ISRA

Consultant
isr,
Brazil,
Joined Apr 2002
3
09:37 May-01-2002
Re: Urgent!!!
I sent this mail 10 days before but i don't know why you don't receive it.

Dear expert
Thanks for giving me of your time to reply my question

by peak number?
i mean the number of half peaks have an amplitude greater than 20% of the
hiegst one (observed by viewing the second backwall echo received from the
V1 block on the oscilloscope)

the kind of probes are broad band normal contact probe (supposed to be
typically constructed)

a third question is
is there any limit for the band width% to judge if it is narrow band or wide
band?





 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1286

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1286
04:10 May-01-2002
Re: Urgent!!!
ISRA:
The number of "peaks" you define is the number of cycles. This varifies my postulation of mechancial variation due to clamping effects on the element.
Smaller dimensioned elements tend to be easier to clamp and thereby limit ring-time. Increased ring-time results in a higher Q-factor [Q=(fc/(f2-f1)] where f2 and f1 are the upper and lower 6dB drop frequencies from the fc or centre frequency.

For a polycrystalline piezo material you will need more damping to reduce the number of cycles over the referenced amplitude for a larger area element as compared to a smaller area element lapped to the same nominal resonant frequency.

As to specific assignments of bandwidth categories; i.e. what percentage bandwidth constitutes broadband or narrowband I do not know what the convention is. I personnally prefer a broadband probe with 1-1.5 cycles and this is usually about a 75-90% bandwidth. But I do not know if a 2 cycled pulse or something less than 70% bandwidth constitutes a narrow band probe.

Band widths greater than 100% are of course possible. But damping rigid crystalline materials to greater and greater amounts eventually kills all useful output. The best option for a useful high bandwidth signal can be had from PVDF and PVTrF materials. These have a very good receiver efficiency (about 7 times more than PZT).

Another potential source of variation is the electrical matching required if there is different capacitance between the electrodes of the large and small probes.
If using a negative spike pulser you can alter the frequency content of the ring by adjusting the pulse duration.

Why are you using the 'second' backwall signal instead of the first? Since you are testing these probes in contact mode you are coupling energy back into the probe's matching plate. Also, because you are testing the frequency content which is, as just discussed above, dependent on the damping of the element you have added another variable in the pressure put on the probe. Now you must consider effects of front loading as well as back loading.

Pages 65-100 of the NDT Handbook Second edition vol 7 by ASNT may be helpful to your design considerations.
Ed

: I sent this mail 10 days before but i don't know why you don't receive it.
.
: Dear expert
: Thanks for giving me of your time to reply my question
.
: by peak number?
: i mean the number of half peaks have an amplitude greater than 20% of the
: hiegst one (observed by viewing the second backwall echo received from the
: V1 block on the oscilloscope)
.
: the kind of probes are broad band normal contact probe (supposed to be
: typically constructed)
.
: a third question is
: is there any limit for the band width% to judge if it is narrow band or wide
: band?
.
.



 
 Reply 
 

Product Spotlight

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Echomac® Small

Available with up to eight channels of electronics to detect and evaluate thickness, flaws and eccen
...
tricity, this UT tester is housed in a convenient, smaller cabinet. This instrument can be used in conjunction with bubbler or immersion tank systems, or with a test bench or in laboratory applications. The Echomac® Small is available in the FD4, FD6 or FD6A versions.
>

Immersion systems

ScanMaster ultrasonic immersion systems are designed for high throughput, multi shift operation in a
...
n industrial or lab environment. These fully integrated systems provide various scanning configurations and incorporate conventional and phased arrays technologies to support diverse applications, such as inspection of disks, bars, shafts, billets and plates. All of ScanMaster immersion systems are built from high accuracy scanning frames allowing for scanning of complex parts and include a multi-channel ultrasonic instrument with exceptional performance. The systems are approved by all major manufacturers for C-scan inspection of jet engine forged discs. Together with a comprehensive set of software modules these flexible series of systems provide the customer with the best price performance solutions.
>

Teletest Focus+

Teletest Focus+ electronics have superior capabilities than rivals on the market. Beyond the usual
...
test features, Focus+ has 24 transmit channels and 24 receive channels, with an additional on-board switching capabilities. The instrument's frequency range is 10–300 kHz.
>

Share...