Helium leaks measurement: ppm or mbar*l/s?

In order to perform valves Fugitive Emissions tests, we use a VARIAN sniffer probe helium leak detector, which gives readings in mbar*l/s. Our usual European reference specification (Shell MESC SPE 77/312) also gives limits in the same measurement unit.
First of all, is anybody able to explain the physical meaning of "mbar*l/s"? Is that a flow rate or a concentration?
Then, how is it possible to relate this with "ppm]"? As many leak detectors manufacturers have instruments with both scales, we think there must be a simple relation between the two units.
Many thanks in advance, 'NGL


 
 Reply 
 
Johnny Darbonne
Johnny Darbonne
00:54 Apr-22-2005
Re: Helium leaks measurement: ppm or mbar*l/s?
: In order to perform valves Fugitive Emissions tests, we use a VARIAN sniffer probe helium leak detector, which gives readings in mbar*l/s. Our usual European reference specification (Shell MESC SPE 77/312) also gives limits in the same measurement unit.
: First of all, is anybody able to explain the physical meaning of "mbar*l/s"? Is that a flow rate or a concentration?
: Then, how is it possible to relate this with "ppm]"? As many leak detectors manufacturers have instruments with both scales, we think there must be a simple relation between the two units.
: Many thanks in advance, 'NGL
.
mbar*l/s is a flow rate of helium into the mass spectrometer. If you know the flow rate on your sniffer probe(measured using a precision flow measuring device) then you can convert your reading to ppm. 1 ppm = 1X10-6 of total flow. i.e. Your probe has a flow rate of 1-mbar-liter/sec and your mass spectrometer reads 5X10-6 mbar-liters/sec helium then the ppm value is 5 ppm helium.


 
 Reply 
 

Product Spotlight

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

PROlineTOP Plug & Play Ultrasonic inspection device

As Plug & Play solution it units all control and operation elements in a small housing and therefo
...
re replaces the typical control cabinet...
>

Conformable wedge transducer

The conformability is obtained with a flexible membrane filled with water between the transducer and
...
the inspected component. The coupling between the membrane and the component requires a small quantity of water or couplant. The conformable wedge combines the acoustic performance of immersion technique with good coupling and low attenuation.
>

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window