13:03 Mar-29-2016
probe output

Dear sir,
I have one doubt ,please help. we have 60 degree probe the prespex shoe was thickness reduced, due to continuous friction . the output of probe wavelength was changed into 65 degree. how can i draw DAC.? for that what is the formula for that..?

 
 Reply 
 
joan pagnoul
Student,
Technifutur, Belgium, Joined Mar 2014, 7

joan pagnoul

Student,
Technifutur,
Belgium,
Joined Mar 2014
7
15:19 Mar-29-2016
Re: probe output
In Reply to vengadesh at 13:03 Mar-29-2016 (Opening).

I think that first, your probe is useless because if I remember norms says 60+/-2 degrees.
And to draw a DAC, methods never change and there is no formula

1
 
 Reply 
 
Csaba Hollo
,
Retired, Canada, Joined Feb 2010, 301

Csaba Hollo

,
Retired,
Canada,
Joined Feb 2010
301
23:29 Mar-29-2016
Re: probe output
In Reply to vengadesh at 13:03 Mar-29-2016 (Opening).

If the code allows you to work with an angle outside of the 60 degree+-, then your calibrations, including regenerating your DAC, must be redone.
Several things have changed, not just your angle, but also your index point and most likely sensitivity.

1
 
 Reply 
 
massimo carminati
Consultant, AUT specialist
IMG Ultrasuoni Srl, Italy, Joined Apr 2007, 691

massimo carminati

Consultant, AUT specialist
IMG Ultrasuoni Srl,
Italy,
Joined Apr 2007
691
19:37 Mar-30-2016
Re: probe output
In Reply to vengadesh at 13:03 Mar-29-2016 (Opening).

Most codes would not allow you to use such a drifted angle from standard. You should recondition the probe or the wedge by adding a new wear sole, but before that you should bring back the original wedge angle to the original value. I warmly not suggest to just use sand paper for doing this.

 
 Reply 
 

Product Spotlight

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

GUL QSR1® Scanning

How do you measure pipe wall thickness without direct access to the area? QSR® Scanning - Guide Wav
...
e Quantitative Short Range Scanning.
>

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

FAAST-PA! OEM Patented phased Array for high speed UT inspection

Multiangle, Multifocus, Multifrequency, Multibeam. Instead of stacking UT electronics and having m
...
any PA probes, FAAST-PA is able to transmit all delay laws within ONE single shot in Real time.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window