probe output

Dear sir,
I have one doubt ,please help. we have 60 degree probe the prespex shoe was thickness reduced, due to continuous friction . the output of probe wavelength was changed into 65 degree. how can i draw DAC.? for that what is the formula for that..?

 
 Reply 
 
joan pagnoul
Student,
Technifutur, Belgium, Joined Mar 2014, 7

joan pagnoul

Student,
Technifutur,
Belgium,
Joined Mar 2014
7
15:19 Mar-29-2016
Re: probe output
In Reply to vengadesh at 13:03 Mar-29-2016 (Opening).

I think that first, your probe is useless because if I remember norms says 60+/-2 degrees.
And to draw a DAC, methods never change and there is no formula

1
 
 Reply 
 
Csaba Hollo
,
Retired, Canada, Joined Feb 2010, 301

Csaba Hollo

,
Retired,
Canada,
Joined Feb 2010
301
23:29 Mar-29-2016
Re: probe output
In Reply to vengadesh at 13:03 Mar-29-2016 (Opening).

If the code allows you to work with an angle outside of the 60 degree+-, then your calibrations, including regenerating your DAC, must be redone.
Several things have changed, not just your angle, but also your index point and most likely sensitivity.

1
 
 Reply 
 
massimo carminati
Consultant, AUT specialist
IMG Ultrasuoni Srl, Italy, Joined Apr 2007, 691

massimo carminati

Consultant, AUT specialist
IMG Ultrasuoni Srl,
Italy,
Joined Apr 2007
691
19:37 Mar-30-2016
Re: probe output
In Reply to vengadesh at 13:03 Mar-29-2016 (Opening).

Most codes would not allow you to use such a drifted angle from standard. You should recondition the probe or the wedge by adding a new wear sole, but before that you should bring back the original wedge angle to the original value. I warmly not suggest to just use sand paper for doing this.

 
 Reply 
 

Product Spotlight

OmniScan™ X3 flaw detector

The OmniScan X3 flaw detector is a complete phased array toolbox. Powerful tools, like total focus
...
ing method (TFM) images and advanced visualization capabilities, enable you to complete your inspection with greater confidence.
>

YXLON Cougar EVO

Scalable small footprint X-ray inspection systems for assembly and laboratory applications. The
...
YXLON Cougar EVO series was designed to provide the "best-in- class" inspection solutions for SMT, semiconductor, and laboratory assembly applications, while maintaining a small system footprint for maximum convenience. With optimized software and hardware, these systems produce higher quality and more consistent results than other electronics inspection systems currently on the market.
>

PROlineTOP Plug & Play Ultrasonic inspection device

As Plug & Play solution it units all control and operation elements in a small housing and therefo
...
re replaces the typical control cabinet...
>

High-end Ultrasonic Flaw Detector with 32:128PR PAUT and 2-ch TOFD: SyncScan 2

SIUI’s newly launched SyncScan 2, is a high-end ultrasonic flaw detector with 32:128PR PAUT and
...
2-ch TOFD, which can maximize your efficiency for PA and TOFD. ● Support PA/TOFD/UT ● 32-ch PA is more suitable for inspection on extra-thick wall and high-attenuation material. ● 32-ch PA and 2-ch TOFD work simultaneously. ● Support PR mode for corrosion inspection.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window