equivalent wire IQI sensitivity

If an ASME plague type penetrameter of a thickness of 2% of the specimen thickness under test is used and the 4T hole is the only hole visible is it right to claim 2% IQI sensitivity. Article 2 of ASME 5 in the appendix section appears to bear this fact up in the wire equivalence table however having talked to various people in the UK they seem to indicate that the 2T hole must be seen even if a plague thickness representing 2% is used. The 4T hole would only indicate a sensitivity of around 2.6%. Can anybody give me further guideance or is this a long standing difference in interpretation between US & European thinking


 
 Reply 
 
Sasaki
Sasaki
01:27 Mar-04-2004
Re: equivalent wire IQI sensitivity
US standards define the acceptance IQI sensitivity in term of e.g. "2-2T" in which "2" stands for the thickness of IQI with regard to the "material" thickness of interest and "2T" denotes "the hole size of the IQI" is what I have been understanding until now.

Regards,


 
 Reply 
 
John O'Brien
Consultant, -
Chevron ETC , USA, Joined Jan 2000, 280

John O'Brien

Consultant, -
Chevron ETC ,
USA,
Joined Jan 2000
280
09:45 Mar-11-2004
Re: equivalent wire IQI sensitivity
You are correct and most Europeans do not read the code properly plus they are hung up on the figure 2% sensitivity which does not appear anywhere in ASME.

ASME is quite clear it tells you you must see a specified hole on a specified thickness. The equivelent penetrameter sensitivity table goes on to say that when you do not have the correct hole IQI you can use the one listed in the EPS Table. Simple no calculations involved.


 
 Reply 
 
ilyes
ilyes
03:27 Oct-27-2004
Re: equivalent wire IQI sensitivity
----------- Start Original Message -----------
:
: Two percent sensitivity is the actual IQI, 2% of the tickness, remember you need to see three sides of the IQI, the 2T hole then is the quality level.
: Exerpt from ASME Section V, Article 22
:
: The quality level usually required for radiography
: is 2% (2-2T when using hole type IQI) unless a higher
: or lower quality is agreed upon between the purchaser
: and the supplier. At the 2% subject contrast level, three quality levels of inspection, 2-1T, 2-2T, and 2-4T, are available through the design and application of the IQI (Practice E 1025, Table 1). Other levels of inspection are available in Practice E 1025, Table 1. The level of inspection specified should be based on the service requirements of the product. Great care should be taken in specifying quality levels 2-1T, 1-1T, and 1-2T byഊSE-94 ARTICLE 22 — RADIOGRAPHIC STANDARDS first determining that these quality levels can be main-tained in production radiography.
:: If an ASME plague type penetrameter of a thickness of 2% of the specimen thickness under test is used and the 4T hole is the only hole visible is it right to claim 2% IQI sensitivity. Article 2 of ASME 5 in the appendix section appears to bear this fact up in the wire equivalence table however having talked to various people in the UK they seem to indicate that the 2T hole must be seen even if a plague thickness representing 2% is used. The 4T hole would only indicate a sensitivity of around 2.6%. Can anybody give me further guideance or is this a long standing difference in interpretation between US & European thinking
------------ End Original Message ------------




 
 Reply 
 
william brennan
william brennan
04:10 Dec-13-2004
IQI Wires and vynal plasic
----------- Start Original Message -----------
: If an ASME plague type penetrameter of a thickness of 2% of the specimen thickness under test is used and the 4T hole is the only hole visible is it right to claim 2% IQI sensitivity. Article 2 of ASME 5 in the appendix section appears to bear this fact up in the wire equivalence table however having talked to various people in the UK they seem to indicate that the 2T hole must be seen even if a plague thickness representing 2% is used. The 4T hole would only indicate a sensitivity of around 2.6%. Can anybody give me further guideance or is this a long standing difference in interpretation between US & European thinking
------------ End Original Message ------------

Can anybody point me in the right direction for finding a supplier of wires,plastic and laminators for making IQI's?



 
 Reply 
 

Product Spotlight

MUSE Mobile Ultrasonic Equipment

The MUSE, a portable ultrasonic imaging system, was developed for in-field inspections of light-weig
...
ht structures. The MUSE consists of a motor-driven manipulator, a water circulation system for the acoustic coupling and a portable ultrasonic flaw detector (USPC 3010). The MUSE provides images of internal defects (A-, B-,C- and D-scan).
>

Navic - Steerable Modular Automated Scanner

The Navic is a modular, motorized, steerable scanner designed to carry multiple attachments used
...
in various scanning and inspection applications. The Navic is capable of weld scanning (girth welds and long seam welds), automated corrosion mapping, and tank scanning.
>

AMIGO2

TSC Amigo2 - ACFM technology has developed a solid reputation for accurately detecting and sizing
...
surface-breaking cracks through paint and coatings. As the industry demands increased performance in speed, signal quality, and portability, it’s time for an evolution. It’s time for Amigo2.
>

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window