11:12 Feb-07-2019
ISO13588 the difference between table A.1 and table A.3

hi, could anyone knows the difference between table A.1 and table A.3, both of them stipulate the length of notch, which one is right? if the thickness of test block is 80mm, what is the length of the notch, 50mm or 40mm?
in my opinion, table A.3 make a mistake, the length of the notch should not be added in table A.3.
could you have any different opinion? thanks

    
 
 
matthias
NDT Inspector, CEO
Dreifeld Materialprüftechnik, Germany, Joined Sep 2005, 81

matthias

NDT Inspector, CEO
Dreifeld Materialprüftechnik,
Germany,
Joined Sep 2005
81
16:08 Feb-07-2019
Re: ISO13588 the difference between table A.1 and table A.3
In Reply to langtuteng at 11:12 Feb-07-2019 (Opening).

To understand this, please look at ISO 10863:2011-12.
They forgot to copy & paste "bottom of the" in Table A.1.

Table A.1 is for one notch at the bottom of the block.
Table A.3 is for three side-drilled holes located at t/4, t/2 and 3t/4 below the surface, with a diameter Dd (see Table A.2) and a length l (Table A.3). Alternatively three notches at the scanning-surface with depths of t/4, t/2 and 3t/4, a tip angle of 60° (see Figure A.5), a width w (Table A.2) and a minimum length of 40 mm may be used.

So for a 80 mm block you need at least one notch:
Notch-bottom, l=50 mm, h= 2 mm.

This is totally not confusing. And good luck with the 0,2 width and 60° tip angle of the notches!

    
 
 
langtuteng
Armenia, Joined Nov 2014, 147

langtuteng

Armenia,
Joined Nov 2014
147
08:22 Feb-09-2019
Re: ISO13588 the difference between table A.1 and table A.3
In Reply to matthias at 16:08 Feb-07-2019 .

thanks very much! most of your answer I couldn't agree with more, except that " Alternatively three notches at the scanning-surface with depths of t/4, t/2 and 3t/4, a tip angle of 60° (see Figure A.5)". according to your opinion, the three notch are surface opening artficial reflector? I think it is just a rectangle reflector which is under the scanning surface 1T/4, 2T/4, 3T/4, not a surface breaking notch and with a tip angle of 60°, and the height、width is based on table A.1
looking forward your opinion. thanks again

    
 
 
Matthias
NDT Inspector, CEO
Dreifeld Materialprüftechnik, Germany, Joined Sep 2005, 81

Matthias

NDT Inspector, CEO
Dreifeld Materialprüftechnik,
Germany,
Joined Sep 2005
81
14:11 Feb-09-2019
Re: ISO13588 the difference between table A.1 and table A.3
In Reply to langtuteng at 08:22 Feb-09-2019 .

Hi, that was not my opinion, I took the text straight out of the standard.

To see if they are top or bottom, check out the drawing

    
 
 
Edward Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1252

Edward Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1252
15:04 Feb-09-2019
Re: ISO13588 the difference between table A.1 and table A.3
In Reply to langtuteng at 08:22 Feb-09-2019 .

Lang, I would point out that Annex A in ISO 13588 is "Informative", not "Normative".
You will read throughout this Annex that the reference reflectors are "recommended".

    
 
 
Udo Schlengermann
Consultant, -
Standards Consulting, Germany, Joined Nov 1998, 176

Udo Schlengermann

Consultant, -
Standards Consulting,
Germany,
Joined Nov 1998
176
18:50 Feb-09-2019
Re: ISO13588 the difference between table A.1 and table A.3
In Reply to Edward Ginzel at 15:04 Feb-09-2019 .

Dear colleagues,

I want to inform, that a revised edition of ISO 13588 has just been published:

ISO 13588:2019-02,
Non-destructive testing of welds - Ultrasonic testing - Use of automated phased array technology.

Annex A of ISO 13588 is still informative:
Annex A
(informative)
Typical reference blocks and reference reflectors.

With best regards
Udo Schlengermann

    
 
 

Product Spotlight

AMIGO2

TSC Amigo2 - ACFM technology has developed a solid reputation for accurately detecting and sizing
...
surface-breaking cracks through paint and coatings. As the industry demands increased performance in speed, signal quality, and portability, it’s time for an evolution. It’s time for Amigo2.
>

MUSE Mobile Ultrasonic Equipment

The MUSE, a portable ultrasonic imaging system, was developed for in-field inspections of light-weig
...
ht structures. The MUSE consists of a motor-driven manipulator, a water circulation system for the acoustic coupling and a portable ultrasonic flaw detector (USPC 3010). The MUSE provides images of internal defects (A-, B-,C- and D-scan).
>

NDT.net launches mobile-friendly design

NDT.net has revamped its website providing a mobile-friendly design.The front page received a comp
...
letely new design and all other sections are now reacting responsively on mobile devices. This has been a major step to make our website more user- friendly.
>

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window