3712 views
Technical Discussions
DEBDUTTA MALLIK
Consultant,
Malaysia, Joined Jun 2005, 37

DEBDUTTA MALLIK

Consultant,
Malaysia,
Joined Jun 2005
37
09:25 Nov-02-2006
TCG CURVE

Dear friends,
Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
Depth of the material: 74mm
Reference reflector: 1.5mm FBH 15mm depth

Your reply will be very useful for me.


Best Regards,


Debdutta Mallik


 
 Reply 
 
Linxiao
Linxiao
08:37 Nov-06-2006
Re: TCG CURVE
Following the link (session 33), you can find several papers on the "2004 ultrasonic benchmark" problem. Their calculations can answer your questions.

good luck!


----------- Start Original Message -----------
: Dear friends,
: Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
: Depth of the material: 74mm
: Reference reflector: 1.5mm FBH 15mm depth
: Your reply will be very useful for me.
:
: Best Regards,
:
: Debdutta Mallik
------------ End Original Message ------------




 
 Reply 
 
Linxiao
Linxiao
08:37 Nov-06-2006
Re: TCG CURVE
Following the link (session 33), you can find several papers on the "2004 ultrasonic benchmark" problem. Their calculations can answer your questions.

http://www.cnde.iastate.edu/QNDE/2004%20Info/Abstract%20Book.pdf

good luck!


----------- Start Original Message -----------
: Dear friends,
: Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
: Depth of the material: 74mm
: Reference reflector: 1.5mm FBH 15mm depth
: Your reply will be very useful for me.
:
: Best Regards,
:
: Debdutta Mallik
------------ End Original Message ------------




 
 Reply 
 
Joe Buckley
Consultant, ASNT L-III, Honorary Secretary of BINDT
Level X NDT, BINDT, United Kingdom, Joined Oct 1999, 528

Joe Buckley

Consultant, ASNT L-III, Honorary Secretary of BINDT
Level X NDT, BINDT,
United Kingdom,
Joined Oct 1999
528
09:44 Nov-06-2006
Re: TCG CURVE
You cannot construct a curve from one point...

However if you have a very good theoretical understanding of the system it might be possible to calculate a curve (in a way similar to the DGS system.) the single reference then allows a sensitivity alignment.

The risk of this is that if you make any incorrect assumptions the curve wont be valid.

Using three or more reference points is the only reliable experimental approach.

Joe


----------- Start Original Message -----------
: Dear friends,
: Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
: Depth of the material: 74mm
: Reference reflector: 1.5mm FBH 15mm depth
: Your reply will be very useful for me.
:
: Best Regards,
:
: Debdutta Mallik
------------ End Original Message ------------




 
 Reply 
 
Ed Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1307

Ed Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1307
02:03 Nov-07-2006
Re: TCG CURVE
If the probe you are using is circular the equations derived by Prof. I. N. Ermolov may be useful to derive an approximate curve. I provided the solutions in a user friendly form in an NDT.net article a few years ago.
http://www.ndt.net/article/v07n01/ginzel/ginzel.htm

By entering the attenuation as 0dB/mm you will get the uncorrected version of the relative amplitudes of the selected reflector. i.e. this is the approximate effect of just the spherical divergence effects.

Ed


----------- Start Original Message -----------
: You cannot construct a curve from one point...
: However if you have a very good theoretical understanding of the system it might be possible to calculate a curve (in a way similar to the DGS system.) the single reference then allows a sensitivity alignment.
: The risk of this is that if you make any incorrect assumptions the curve wont be valid.
: Using three or more reference points is the only reliable experimental approach.
: Joe
:
: : Dear friends,
: : Please let me know the method to draw a TCG curve in Ultrasonic testing system when there is only one reference reflector. Is there any calculation on the basis of depth of the material,amplitude or the size of reflector.What will be the rate of decreasing in dB upto zero so that I become same sensitive at all depth.Following are the informatiopn for your understanding.
: : Depth of the material: 74mm
: : Reference reflector: 1.5mm FBH 15mm depth
: : Your reply will be very useful for me.
: :
: : Best Regards,
: :
: : Debdutta Mallik
------------ End Original Message ------------




 
 Reply 
 
Khurath
SIS, India, Joined Jan 2018, 2

Khurath

SIS,
India,
Joined Jan 2018
2
06:25 Mar-14-2019
Re: TCG CURVE
In Reply to Joe Buckley at 09:44 Nov-06-2006 .

Thanks for this information of sensitivity calibration and for above 74mm thickness what should be our material debt for calibration

 
 Reply 
 

Product Spotlight

Scatter Correction for Industrial Cone-Beam Computed Tomography (CBCT)

Download our white paper on Scatter Correction for Industrial Cone-Beam Computed Tomography (CBCT)
...
Using 3D VSHARP, a fast GPU-Based Linear Boltzmann Transport Equation Solver here: http://bit.ly/2Y0t78P
>

Extended Range Variable Wall® Piping Calibration Blocks for Longseam Inspection

This set of two blocks has been specially designed for situations where ultrasonic examination tak
...
es place circumferentially or “around the curve” for applications such as inspection of longseam welds in piping and vessels. Jointly designed by PH Tool and Holloway NDT, the blocks feature side-drilled hole reflectors positioned within novel compound-curves, each with multiple thickness steps. The unique design results in a light and compact form factor, with the largest block weighing just over 15 lbs and spanning just 16 ¾”. Coverage is provided for nominal pipe sizes from 6” to 10” in the small block and 12” to a whopping 56” in the large block. The compound OD curves and multiple wall thicknesses actually exceed ASME requirements, providing exceptional signal response and measurement accuracy. Properly managing beam divergence off the ID and stabilizing beam skew due to probe rocking is essential to maintaining a consistent calibration when inspecting lengthwise on curved surfaces.
>

Micro CT

nanoVoxel series has nano-scale resolution upto 500nm by using high voltage X-ray source (from 80-
...
300 kV) with micro/nano focal spot and highly sensitive flat panel detectors and unique Optical coupling detector.
>

TESTD-PT SYSTEM