21:06 Apr-13-2019
PAUT DATA NOISE

Greetings to everyone
I have a doubt regarding paut data
Today when I collect my PAUT data in a pipeline ,there was a lot of electrical noise like signals coming.In C scan it's full of doted pixels.There was no welding on that line and no other hot works.Electrically also isolated.I was using omniscan mx2 .Can anyone tell me from where it's coming and how can I solve this issue?

Thanks.

    
 
 
Anmol Birring
Consultant,
Birring NDE Center, Inc., USA, Joined Aug 2011, 679

Anmol Birring

Consultant,
Birring NDE Center, Inc.,
USA,
Joined Aug 2011
679
05:00 Apr-14-2019
Re: PAUT DATA NOISE
In Reply to Noufel at 21:06 Apr-13-2019 (Opening).

Check the scanner motors
All equipment should be connected to common ground

    
 
 
Ali
NDT Inspector,
Iran, Joined Mar 2013, 82

Ali

NDT Inspector,
Iran,
Joined Mar 2013
82
10:13 Apr-14-2019
Re: PAUT DATA NOISE
In Reply to Noufel at 21:06 Apr-13-2019 (Opening).

Do you have same noise on the cal blocks? Is the noise happen before start the scan or only it comes during scaning? Is the power chargers connect to equipment ?

    
 
 
Matthias
NDT Inspector, CEO
Dreifeld Materialprüftechnik, Germany, Joined Sep 2005, 79

Matthias

NDT Inspector, CEO
Dreifeld Materialprüftechnik,
Germany,
Joined Sep 2005
79
17:44 Apr-14-2019
Re: PAUT DATA NOISE
In Reply to Noufel at 21:06 Apr-13-2019 (Opening).

I had exactly the same, it looked like indications flying through the sector scan.
The solution was to use a filter (10MHz low pass on this case).

    
 
 
Nick Bublitz
Other, Phased Array Analysist
VeriPhase, USA, Joined May 2010, 122

Nick Bublitz

Other, Phased Array Analysist
VeriPhase,
USA,
Joined May 2010
122
14:53 Apr-15-2019
Re: PAUT DATA NOISE
In Reply to Noufel at 21:06 Apr-13-2019 (Opening).

A likely cause is too high of PRF/Scanning speed. Optimize your setup (range, gates etc.) or lower PRF/Scanning speed.

    
 
 
Noufel
McDERMOTT MIDDLE EAST INC-DUBAI, United Arab Emirates (UAE), Joined Sep 2018, 7

Noufel

McDERMOTT MIDDLE EAST INC-DUBAI,
United Arab Emirates (UAE),
Joined Sep 2018
7
18:51 Apr-18-2019
Re: PAUT DATA NOISE
In Reply to Anmol Birring at 05:00 Apr-14-2019 .

I am using hand scan (multimag from phoenix )

    
 
 
Noufel
McDERMOTT MIDDLE EAST INC-DUBAI, United Arab Emirates (UAE), Joined Sep 2018, 7

Noufel

McDERMOTT MIDDLE EAST INC-DUBAI,
United Arab Emirates (UAE),
Joined Sep 2018
7
19:00 Apr-18-2019
Re: PAUT DATA NOISE
In Reply to Ali at 10:13 Apr-14-2019 .

No outside power connected to the system except omniscan scans battery

    
 
 
Noufel
McDERMOTT MIDDLE EAST INC-DUBAI, United Arab Emirates (UAE), Joined Sep 2018, 7

Noufel

McDERMOTT MIDDLE EAST INC-DUBAI,
United Arab Emirates (UAE),
Joined Sep 2018
7
19:04 Apr-18-2019
Re: PAUT DATA NOISE
In Reply to Matthias at 17:44 Apr-14-2019 .

I am using 5L64A12 probe with filter none.
As you said can we use 10 MHZ filter for this probe?
Is there any criteria for choosing filter for a particular frequency?

    
 
 
Noufel
McDERMOTT MIDDLE EAST INC-DUBAI, United Arab Emirates (UAE), Joined Sep 2018, 7

Noufel

McDERMOTT MIDDLE EAST INC-DUBAI,
United Arab Emirates (UAE),
Joined Sep 2018
7
19:08 Apr-18-2019
Re: PAUT DATA NOISE
In Reply to Nick Bublitz at 14:53 Apr-15-2019 .

PRF is already at minimum as possible.and setup optimised for range and gates ,still noise is present.
I am using linear scan with aperture 16 for 4 groups also having tofd group.

    
 
 
Matthias
NDT Inspector, CEO
Dreifeld Materialprüftechnik, Germany, Joined Sep 2005, 79

Matthias

NDT Inspector, CEO
Dreifeld Materialprüftechnik,
Germany,
Joined Sep 2005
79
22:24 Apr-19-2019
Re: PAUT DATA NOISE
In Reply to Noufel at 19:04 Apr-18-2019 .

Choose the filter before calibration and pick whatever filter gives you the best signal.

    
 
 
Pratik Wagh
Consultant,
NDT Video Library on YouTube, USA, Joined Oct 2011, 16

Pratik Wagh

Consultant,
NDT Video Library on YouTube,
USA,
Joined Oct 2011
16
18:41 Apr-20-2019
Re: PAUT DATA NOISE
In Reply to Noufel at 19:08 Apr-18-2019 .

Have you tried Scan>>Scan Speed . Lower it till the noise disappears. If nothing works, sometimes it is just best to restart with a fresh new setup and it will magically disappear. Or if you figure out how you did, please share it here.

    
 
 
Charles Hansen
, ASNT Level III Technical Coordinator
USA, Joined May 2018, 1

Charles Hansen

, ASNT Level III Technical Coordinator
USA,
Joined May 2018
1
16:19 Apr-22-2019
Re: PAUT DATA NOISE
In Reply to Noufel at 19:08 Apr-18-2019 .

Noufel,

Can you upload your data?

    
 
 

Product Spotlight

AMIGO2

TSC Amigo2 - ACFM technology has developed a solid reputation for accurately detecting and sizing
...
surface-breaking cracks through paint and coatings. As the industry demands increased performance in speed, signal quality, and portability, it’s time for an evolution. It’s time for Amigo2.
>

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

NEW - TD Focus-ScanRX

The NEW Next Generation Advanced UT platform, TD Focus ScanRX - Also available as a card stack solut
...
ion. Key Improvements 1. Data acquisition is significantly faster than current design 2. Better aesthetic – closely aligns with HandyScan RX 3. Improved IP rating (Target IP66) 4. Ruggedized housing 5. Connectors are protected from impact and ingress 6. Integrated stand and separate retractable handle easy to keep clean) 7. Touchscreen with ruggedized display glass 8. 3-Axis encoder input
>

FlexoFORM™

The FlexoFORM™ solution integrates flexible phased array probe technology in a scanner to solv
...
e the challenges that come with inspecting pipe elbows. When combined with a water column, this solution can be used to collect easy-to-interpret data on elbows with diameters ranging from 4.5 in. OD up to flat and offers many benefits.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window