Mark F.
Mark F.
00:55 May-14-2019
Olympus 38DL Plus Question

Hello all,

I recently acquired a 38DL Plus from Olympus and am currently attempting to measure the thickness of the 303 Stainless Reference (P/N: 2218E) that came with the instrument utilizing a M202-RM Delay Line Single Transducer (10 MHz). However I can not for the life of me get anything other than a "LOS" measurement and when calibrating it says the calibration fails each time.

Thoughts? For reference I am using Olympus Couplant-B (P/N: B-2)

Regards,

Mark

    
 
 Reply 
 
SUDHEER JAI KRISHNAN
Other,
Olympus, United Arab Emirates (UAE), Joined Sep 2018, 21

SUDHEER JAI KRISHNAN

Other,
Olympus,
United Arab Emirates (UAE),
Joined Sep 2018
21
10:33 May-14-2019
Re: Olympus 38DL Plus Question
In Reply to Mark F. at 00:55 May-14-2019 (Opening).

Hi Mark

Sorry to ask this question, but need to know whether you have applied the couplant between the delay and the transducer?

    
 
 Reply 
 
Mark F.
Mark F.
10:47 May-14-2019
Re: Olympus 38DL Plus Question
In Reply to SUDHEER JAI KRISHNAN at 10:33 May-14-2019 .

Between the delay and the transducer? Is it not supposed to be placed on the calibration piece itself?

    
 
 Reply 
 
Sudheer Jai Krishnan
Other,
Olympus, United Arab Emirates (UAE), Joined Sep 2018, 21

Sudheer Jai Krishnan

Other,
Olympus,
United Arab Emirates (UAE),
Joined Sep 2018
21
10:52 May-14-2019
Re: Olympus 38DL Plus Question
In Reply to Mark F. at 10:47 May-14-2019 .

You have to apply couplant between the delay line and the transducer in order for the sound to pass through the delay .

After that , you can take your measurements as usual , by placing the couplant on the test block

Let us know if it works

1    
 
 Reply 
 
Mark F.
Mark F.
21:45 May-14-2019
Re: Olympus 38DL Plus Question
In Reply to Sudheer Jai Krishnan at 10:52 May-14-2019 .

Is there any need to calibrate the device before taking a measurement? Every time I try to calibrate a thick and thin piece it fails with "Calibration value invalid".

    
 
 Reply 
 
ASHOK KUMAR PATRA
,
Saudi K-KEM Engineering service co, Saudi Arabia, Joined May 2018, 52

ASHOK KUMAR PATRA

,
Saudi K-KEM Engineering service co,
Saudi Arabia,
Joined May 2018
52
10:44 May-16-2019
Re: Olympus 38DL Plus Question
In Reply to Mark F. at 21:45 May-14-2019 .

Dear Mark,

Please open the probe delay block and apply some Couplant and again fix it.

Then you fix the probe and select the probe XDCR recall and single line transducer .

Then press 2nd F and Cal Zero.
Is this giving you any zero value. If it is giving some value is very good.

Calibrate now. you will definitely able to calibrate

    
 
 Reply 
 

Product Spotlight

Compact NDT inspection-heads for measurements with active thermography

The compact inspection head is suitable for thermographic ndt tasks. The uncooled infrared camera
...
is specially developed for NDI-tasks and offers a thermal sensitivity until now known only from thermal imagers with cooled detector. All required components and functions are integrated into the inspection-head. You will only need an ethernet cable to connect the sensor with the evaluation system.
>

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

GUL QSR1® Scanning

How do you measure pipe wall thickness without direct access to the area? QSR® Scanning - Guide Wav
...
e Quantitative Short Range Scanning.
>

PROlineTOP Plug & Play Ultrasonic inspection device

As Plug & Play solution it units all control and operation elements in a small housing and therefo
...
re replaces the typical control cabinet...
>