Mark F.
Mark F.
00:55 May-14-2019
Olympus 38DL Plus Question

Hello all,

I recently acquired a 38DL Plus from Olympus and am currently attempting to measure the thickness of the 303 Stainless Reference (P/N: 2218E) that came with the instrument utilizing a M202-RM Delay Line Single Transducer (10 MHz). However I can not for the life of me get anything other than a "LOS" measurement and when calibrating it says the calibration fails each time.

Thoughts? For reference I am using Olympus Couplant-B (P/N: B-2)

Regards,

Mark

    
 
 Reply 
 
SUDHEER JAI KRISHNAN
Other,
Olympus, United Arab Emirates (UAE), Joined Sep 2018, 21

SUDHEER JAI KRISHNAN

Other,
Olympus,
United Arab Emirates (UAE),
Joined Sep 2018
21
10:33 May-14-2019
Re: Olympus 38DL Plus Question
In Reply to Mark F. at 00:55 May-14-2019 (Opening).

Hi Mark

Sorry to ask this question, but need to know whether you have applied the couplant between the delay and the transducer?

    
 
 Reply 
 
Mark F.
Mark F.
10:47 May-14-2019
Re: Olympus 38DL Plus Question
In Reply to SUDHEER JAI KRISHNAN at 10:33 May-14-2019 .

Between the delay and the transducer? Is it not supposed to be placed on the calibration piece itself?

    
 
 Reply 
 
Sudheer Jai Krishnan
Other,
Olympus, United Arab Emirates (UAE), Joined Sep 2018, 21

Sudheer Jai Krishnan

Other,
Olympus,
United Arab Emirates (UAE),
Joined Sep 2018
21
10:52 May-14-2019
Re: Olympus 38DL Plus Question
In Reply to Mark F. at 10:47 May-14-2019 .

You have to apply couplant between the delay line and the transducer in order for the sound to pass through the delay .

After that , you can take your measurements as usual , by placing the couplant on the test block

Let us know if it works

1    
 
 Reply 
 
Mark F.
Mark F.
21:45 May-14-2019
Re: Olympus 38DL Plus Question
In Reply to Sudheer Jai Krishnan at 10:52 May-14-2019 .

Is there any need to calibrate the device before taking a measurement? Every time I try to calibrate a thick and thin piece it fails with "Calibration value invalid".

    
 
 Reply 
 
ASHOK KUMAR PATRA
,
Saudi K-KEM Engineering service co, Saudi Arabia, Joined May 2018, 52

ASHOK KUMAR PATRA

,
Saudi K-KEM Engineering service co,
Saudi Arabia,
Joined May 2018
52
10:44 May-16-2019
Re: Olympus 38DL Plus Question
In Reply to Mark F. at 21:45 May-14-2019 .

Dear Mark,

Please open the probe delay block and apply some Couplant and again fix it.

Then you fix the probe and select the probe XDCR recall and single line transducer .

Then press 2nd F and Cal Zero.
Is this giving you any zero value. If it is giving some value is very good.

Calibrate now. you will definitely able to calibrate

    
 
 Reply 
 

Product Spotlight

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

IRIS 9000Plus - Introducing the next generation of heat exchanger inspection.

Representing the seventh generation of the IRIS system, the IRIS 9000 Plus has nearly 200 years of c
...
ombined field inspection experience incorporated in its design. This experience combined with a strong commitment to quality and a history of innovation has made Iris Inspection Services® the undisputed leader in IRIS technology.
>

iProbe - USB Phased Array Probe

Turn your PC, Laptop or Tablet into a powerful 32:64 Phased Array system with our USB powered Phas
...
ed Array iProbe and inspection software package. Perform affordable weld inspection and corrosion mapping inspections with this powerful but small system. The probe has a 100V pulser, 8K PRF, dual axis encoder input and achieves outstanding signal quality.
>

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window