18:40 May-18-2019
Why sizing @ 20% and below FSH is not recommended

Need a technical write up, thought process to confirm
Mathematical approach, to explain.

Secondly in other words DAC/ DGS points below 20%. Are converted
To split DAC/ DGS. Why? And the why its recommended?

Regards
Pramod Myakal
RSM Asia
Sonatest LTD UK.

    
 
 
Paul Holloway
Consultant, UT Level 3
Holloway NDT & Engineering Inc , Canada, Joined Apr 2010, 169

Paul Holloway

Consultant, UT Level 3
Holloway NDT & Engineering Inc ,
Canada,
Joined Apr 2010
169
21:01 May-18-2019
Re: Why sizing @ 20% and below FSH is not recommended
In Reply to Pramod at 18:40 May-18-2019 (Opening).

Because it's hard to see things that are only 20% screen height.

    
 
 
Martin Goodman
,
United Kingdom, Joined Apr 2019, 4

Martin Goodman

,
United Kingdom,
Joined Apr 2019
4
07:30 May-19-2019
Re: Why sizing @ 20% and below FSH is not recommended
In Reply to Paul Holloway at 21:01 May-18-2019 .

Sizing at 20% FSH is generally for height sizing of linear defects. Length sizing will normally be done at 50% (6db drop). The reason for 20% sizing is that more of the beam is on the defect than for length sizing where half the beam is on the defect. If you use LOG20x100/20 you will see your Db difference between FSH and 20%.

    
 
 
Anmol Birring
Consultant,
Birring NDE Center, Inc., USA, Joined Aug 2011, 701

Anmol Birring

Consultant,
Birring NDE Center, Inc.,
USA,
Joined Aug 2011
701
19:34 May-19-2019
Re: Why sizing @ 20% and below FSH is not recommended
In Reply to Pramod at 18:40 May-18-2019 (Opening).

Results in oversizing

    
 
 
Edward Ginzel
R & D, -
Materials Research Institute, Canada, Joined Nov 1998, 1252

Edward Ginzel

R & D, -
Materials Research Institute,
Canada,
Joined Nov 1998
1252
20:31 May-19-2019
Re: Why sizing @ 20% and below FSH is not recommended
In Reply to Pramod at 18:40 May-18-2019 (Opening).

Pramod, you have been getting a few different perspectives on this, but that is probably because it is not clear what you are asking about. Your second statement begins with "Secondly in other words..." but you did not make a first question, just a statement that you need to explain "something".
There are 2 aspects in the second paragraph you made. One relates to DAC/DGS. I think the DGS is not correct as we usually see the curves for the various diameter DSRs drawn down to the horizontal baseline. But a "Split DAC" is usually split because, as Paul noted, things are difficult to see at such a low amplitude. A DAC point at 80% screen height allows you to see a significant change in signal amplitude to represent 2dB (e.g. 80% to 63% represents 2dB but a 20% DAC point allows only 4% change in screen height before you have reached 2dB).

The title you placed on the thread seems to ask about sizing (not Split DACs). This SEEMS to suggest you are now asking about the probe movement to assess flaw size (length or height). This is not the same rationale for a Split DAC. Using the -6dB, -14dB or -20dB drop from a maximum or from the DAC level are merely options for the empirical beam-boundary sizing techniques. For sizing laminations that are larger than the beam dimension the planar nature of the flaw makes the -6dB option a reasonable estimate of the extents. But when trying to estimate the length of a nonplanar flaw (e.g. cluster porosity or slag), the tenuous nature of the flaw will provide lower amplitude responses at the ends of the flaw extent. Therefore even though you have moved the probe such that the amplitude has dropped to -6dB, it will probably not be a suitable indication of the end of the flaw. For this reason, it is sometimes recommended to use a -14 or -20dB boundary to indicate the flaw extent.

    
 
 
Paul Holloway
Consultant, UT Level 3
Holloway NDT & Engineering Inc , Canada, Joined Apr 2010, 169

Paul Holloway

Consultant, UT Level 3
Holloway NDT & Engineering Inc ,
Canada,
Joined Apr 2010
169
20:39 May-19-2019
Re: Why sizing @ 20% and below FSH is not recommended
In Reply to Martin Goodman at 07:30 May-19-2019 .

I'm not so crazy about the 6dB drop method for length sizing. It has a high probability of undersizing flaws.

Article here: http://www.hollowayndt.com/news/2018/10/22/the-shortcomings-of-the-6db-drop-method-for-ultrasonic-flaw-length-sizing

    
 
 
Martin Goodman
,
United Kingdom, Joined Apr 2019, 4

Martin Goodman

,
United Kingdom,
Joined Apr 2019
4
10:58 May-21-2019
Re: Why sizing @ 20% and below FSH is not recommended
In Reply to Paul Holloway at 20:39 May-19-2019 .

I completely agree Paul, and that's a good article explaining the pit-falls of sizing. I started out using 20 / 6 Db drops for sizing, and it is still widely used as a standard technique.

6Db makes much more sense for laminations obviously.

1    
 
 
Ricardo Andreucci
FATEC - Technological University of Sao Paulo, Brazil, Joined May 2018, 17

Ricardo Andreucci

FATEC - Technological University of Sao Paulo,
Brazil,
Joined May 2018
17
17:58 May-27-2019
Re: Why sizing @ 20% and below FSH is not recommended
In Reply to Paul Holloway at 20:39 May-19-2019 .

Hi Paul,
I have read the article "the short comings of the 6db drop method for ultrasonic flaw length sizing", by the way is very interesting, but I believe that most of the time repairing rejected indications in welds detected by ultrasonic testing, are performed in dimensions larger than the technician sized. In addition the repair cavity normally is performed liquid penetrant testing in order to ensure the elimination of the defect.
So, in my opinion I think that sizing by 20dB or 6dB drop for rejected descontinuities whatever for me.

Regards

    
 
 

Product Spotlight

NovaScope 6000

The all-digital Novascope 6000 is a portable, ultra-high precision thickness gauge for high-speed
...
thickness measurement. Novascope 6000 has unmatched capabilities and unique features including: •Superior Resolution with high contrast, high-speed color RF display •High pulser voltage •Real-time video output •Increased internal/external data storage •Programmable SetUp features •Battery & AC Powered
>

TOPAZ® 64: Fully Integrated Portable 64 Channel PAUT Instrument with FMC & TFM

Introducing TOPAZ64, the industry’s most intelligent fully integrated, portable 64-channel phased
...
array ultrasound (UT) instrument. TOPAZ64 combines code-compliant phased array UT with the industry’s most advanced full matrix capture (FMC) and total focusing method (TFM) capabilities. Featuring the highest acquisition frequency in its class, high resolution FMC and a 12” multi-touchscreen, users can easily visualize even the smallest flaws. TOPAZ64 can generate a bipolar pulse that provides more acoustic energy versus previous models for punching through thick components. The result is a portable tool that delivers increased inspection coverage, more accurate signals, and the ability to handle all UT inspections in one package. TOPAZ64 is ideal for challenging applications in transportation, oil and gas, manufacturing and power generation.
>

Swift & Scorpion2 - a state-of-the-art remote-control ultrasonic crawler

The Swift and Scorpion2 dry-coupled, remote-access ultrasonic crawler bring major efficiency and d
...
ata improvements to tank shell inspections and other structures such as vessels and offshore installations.
>

SONOAIR - air-coupled Phased Array Ultrasonic Inspection System

For highly attenuating materials, the performance of the system is critical. The ultrasonic sensors,
...
the scanning area and the system settings should be flexibly adapted to the test task and the material. These high expectations are met with the new and modular testing system SONOAIR. With the world’s first air-coupled phased-array UT inspection system SONOAIR we developed a technology that works with up to 4 transmitter and receiver channels with freely configurable square wave burst transmitters as wells as low noise receiving amplifiers.
>