I have some small doubts in UT
Download

1. There are calibration v2 blocks of 1.5 hole and 5 me hole what is the reason behind it is it only for better sensitivity or there were any other reson

2. While doing range calibration with v2 block I observed these things with Einstein modsonic 2 that while we faced the probe from 25 radius end of we clearly observe the depth it will be shown as 8.5 but where we given the thickness of the material as more than 8.5 but in the same case if the thickness is below it changes the value either 2 to 4.5 . Why?

3.what is the weld geometry for a pipe especially for double v saw weld
Pipe,

4. Is it really necessary to add the reinforcement of I'd and od to thickness of the parent material in the case where we using N5 notches (example coil thickness 15 and reinforcement is 1.5+1.5 so is iit mandetary to add 3mm to job thickness or not) .

5.can anyone share me the actual cycle time calculation ( I have doubt of taking the dia of the crystal as 8 or 12 for (8*9 rectangular crystprobe)

For 5 one just tell me what is correct. As we are facing 8' crystal and doing so we need to consider only 8 or as a formula we need to use 12 .
Just see attached link for my reference material and find page 449-56 page https://inis.iaea.org › _PublicPDF
Web results
ultrasonic testing of materials at level 2 - International Nuclear Information

https://www.google.com/url?sa=t&source=web&rct=j&url=https://inis.iaea.org/collection/NCLCollectionStore/_Public/19/100/19100874.pdf&ved=2ahUKEwi4lemBpJnkAhXVgeYKHUPGA7IQFjAAegQIAxAB&usg=AOvVaw33fQxSLXw0tnOFXLgU09kv

 
 Reply 
 
Rahul
India, Joined Jul 2019, 6

Rahul

India,
Joined Jul 2019
6
20:27 Aug-23-2019
Re: I have some small doubts in UT
In Reply to Chandrasekhar at 17:15 Aug-23-2019 (Opening).

Hi,
Answer for your second question..
When thickness is 8.5 or above, it shows tk as 8.5 means sound waves actually reflect at that point.
When it's below 8.5, tk will get reduce.. as machine trigonometry thinks the waves passing in the full skip which is not happen.
Here in m/c set tk as 7. Now waves after 7mm tk will be considered as it travels in full skip. So you will get tk 5.5.


 
 Reply 
 

Product Spotlight

NOVO Armor 15 & NOVO Armor 22

The Armor Kit Contains the NOVO Armor, which provides additional mechanical protection to the NOVO 1
...
5WN & NOVO 22WN Detectors, the Armor Stand and a traveling soft cover. - Newest shock absorbent technology case - Water resistant design - Supports wired & wireless communication - Multiple positioning options - Tripod connection using the Built-in 1/4” threads - Simple Detector battery replacement
>

NDTkit RT

NDTkit RT, TESTIA's Digital Radiography software The NDTkit product line software for X-ray analysi
...
s. NDTkit RT is a software benefiting from the Ultis kernel which is dedicated to radiographic image analysis. It offers a set of tools and filtering processes to assist RT operators in finding relevant flaws.
>

iProbe - USB Phased Array Probe

Turn your PC, Laptop or Tablet into a powerful 32:64 Phased Array system with our USB powered Phas
...
ed Array iProbe and inspection software package. Perform affordable weld inspection and corrosion mapping inspections with this powerful but small system. The probe has a 100V pulser, 8K PRF, dual axis encoder input and achieves outstanding signal quality.
>

Sci Aps Z-Series Portable Handheld Analysers

The world’s only handheld analyzer that measures carbon content in stainless (yes even L-grades),s
...
teels, and cast irons. Also accepted for low Si analysis for sulfidic corrosion analysis, and is widely used in the power industry for Cr analysis, for flow accelerated corrosion applications.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window