06:34 Sep-09-2019
How to select Rectifier in Ultrasonic testing.

Selection of Rectifier in ultrasonic testing ,like full wave,half wave positive ,half waves negatives or radio frequency ,what is criteria selection of that mainly doing operation on carbon composite material. We are using Omni scan sx by olympus.

 
 Reply 
 
nick bublitz
Other, Phased Array Analysist
VeriPhase, USA, Joined May 2010, 132

nick bublitz

Other, Phased Array Analysist
VeriPhase,
USA,
Joined May 2010
132
15:04 Sep-09-2019
Re: How to select Rectifier in Ultrasonic testing.
In Reply to Lirin Philip at 06:34 Sep-09-2019 (Opening).

Generically-
Full wave for angle beam flaw detection where maximum echo response is of most importance. RF to visualize phase shift- bond testing etc. Half wave for thickness testing/corrosion testing.
You can search for and apply conventional ut principles on these displays to phased array.

from Olympus tutorial on UT:
Received echoes can be displayed either as an unrectified RF signal, as half wave positive or half wave negative rectified signals, or with full wave rectification. Raw echoes are initially processed as RF waveforms with both positive and negative peaks. The RF display mode is useful when working with very thin test pieces and in cases where echo phase or polarity is of interest. Half wave positive rectification shows only the positive peaks, while half wave negative rectification shows only the negative peaks, flipped to the positive side of the baseline. Half wave rectification can in some cases improve signal-to-noise by reducing spurious peaks. Full wave rectification displays both positive and negative echo lobes together on the positive side of the baseline and is the most commonly used format in flaw detection applications

3
 
 Reply 
 
Paul Holloway
Consultant,
Holloway NDT & Engineering Inc , Canada, Joined Apr 2010, 233

Paul Holloway

Consultant,
Holloway NDT & Engineering Inc ,
Canada,
Joined Apr 2010
233
15:18 Sep-09-2019
Re: How to select Rectifier in Ultrasonic testing.
In Reply to Lirin Philip at 06:34 Sep-09-2019 (Opening).

zoom image



Further to Nick's comments, half-wave is best used in thickness testing to ensure that measurements are always made on wave lobes of the same polarity.

If you use full wave rectification, it is impossible to tell whether your gate is triggering on a positive or negative lobe (see attached image). This may be inconsequential if your measurement resolution requirements are relatively low (i.e. if you don't care about the nearest 0.005" or 0.010", then full wave may be fine!).

But if you are doing very fine, precise work (e.g. pen probe or small diameter high frequency dually) then you are definitely better off using half wave.

Use RF mode to start, find out which wave lobe is the best, then calibrate off that and go for it.
2
 
 Reply 
 
Lirin Philip
,
VALDEL , India, Joined Sep 2019, 10

Lirin Philip

,
VALDEL ,
India,
Joined Sep 2019
10
10:17 Sep-10-2019
Re: How to select Rectifier in Ultrasonic testing.
In Reply to Paul Holloway at 15:18 Sep-09-2019 .

I'm working in the field of composite materials and the maximum thickness of composite parts comes around less than 5mm.

 
 Reply 
 

Product Spotlight

Combination of Digital Image Correlation and Thermographic Measurements

The combination of measuring results from the digital image correlation (ARAMIS, DIC) and temperat
...
ure measuring data from infrared cameras permits the simultaneous analysis of the thermal and mechanical behavior of test specimens in the materials and components testing field.
>

On-Demand Webinar: Advanced Ultrasonic Techniques for HTHA Detection & Assessment

In the Oil and Gas industry, High Temperature Hydrogen Attack (HTHA) can occur in steels operating a
...
t high temperature, usually above 205°C (400°F). Early stages of HTHA are very difficult to detect. Join us for a webinar presented by Guy Maes, Zetec Sales Engineer Director. The webinar will demonstrate how advanced ultrasonic techniques are deployed for efficient and reliable HTHA inspections.
>

IntraPhase Athena Phased Array System

The Athena Phased Array system, manufactured by WesDyne NDE Products & Technology, consists of a pha
...
sed array acquisition system and PC running IntraSpect software. A PC is used to perform acquisition, analysis and storage of the data. System hardware is capable of operating up to four data sets with any combination of phased array or conventional UT probes. NOW AVAILABLE IN 64-64 CONFIGURATION.
>

PAUT Probes

Typical Phased Array probes have frequencies between 1MHz and 20MHz and the number of wafers is 10
...
to 128. M2 Electronics offers customers conventionally ultrasound probes and the ability to provide high-precision Phased Array Ultrasound Probes of up to 256 wafers. We can also customize the probe for our customers to meet the specific application requirements of the user.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window