Influence of pulse frequency on measurement of material thickness change

Hi guys,

I'm a bit confused with the following question. Assuming the signal sampling frequency is sufficiently high, how the ultrasonic pulse frequency influences the measurement precision of material thickness change in a pulse-echo mode?

The example of the problem is this. Material with c=3.4 mm/microsec is grinded by a 0.1 mm steps from top to bottom, signal sampling frequency is very high (1.25 GHz) and the frontwall and backwall echoes are clearly visible. What is the basic difference if a transducer with 1MHz, 2.25 or 3.5 MHz or higher is used for that?

 
 Reply 
 
Anmol Birring
Consultant,
Birring NDE Center, Inc., USA, Joined Aug 2011, 832

Anmol Birring

Consultant,
Birring NDE Center, Inc.,
USA,
Joined Aug 2011
832
11:02 Dec-11-2019
Re: Influence of pulse frequency on measurement of material thickness change
In Reply to Alex at 08:35 Dec-11-2019 (Opening).

No difference, but you got to use zero crossing method.

 
 Reply 
 
Wiesław Bicz
Engineering,
PBP Optel sp. z o.o., Poland, Joined Feb 2009, 276

Wiesław Bicz

Engineering,
PBP Optel sp. z o.o.,
Poland,
Joined Feb 2009
276
12:38 Dec-11-2019
Re: Influence of pulse frequency on measurement of material thickness change
In Reply to Alex at 08:35 Dec-11-2019 (Opening).

Time of flight measurement has not much to do with frequency, but with the method of signal analysis. This was already discussed on this forum many times. Frequency of signal is of much lower importance than for example the accuracy of amplitude measurement.
Zero crossing method mentioned by Anmol is one of the simplest possibilities.

 
 Reply 
 
Alex
,
Israel, Joined Sep 2017, 8

Alex

,
Israel,
Joined Sep 2017
8
14:16 Dec-11-2019
Re: Influence of pulse frequency on measurement of material thickness change
In Reply to Wiesław Bicz at 12:38 Dec-11-2019 .

Hi Anmol and Wieslaw, thank you for your response.
I use envelope extraction and cross-correlation to estimate the time of flight. And use this method consistently between all acquired signals on the course of material thickness change.

Can you though, describe why the pulse wavelength has nothing to do with the distance change measurement? Can I measure smaller and smaller increments of the thickness change only by carrying for sufficiently high sampling period?

 
 Reply 
 
Wiesław Bicz
Engineering,
PBP Optel sp. z o.o., Poland, Joined Feb 2009, 276

Wiesław Bicz

Engineering,
PBP Optel sp. z o.o.,
Poland,
Joined Feb 2009
276
18:08 Dec-12-2019
Re: Influence of pulse frequency on measurement of material thickness change
In Reply to Alex at 14:16 Dec-11-2019 .

There are two critical factors in TOF measurement:
The most important factor is the accuracy of amplitude measurement.
In the case if you want to measure the distance to some pulse measured from the excitation moment, the synchronization between the pulser excitation and sampling clock is also critical.
In the case of measurement between two pulses practically only the accuracy of amplitude measurement. Sampling rate must be sufficient from the point of view of Nyquist criterion, but is of significantly lower importance than the amplitude measurement accuracy.
And naturally the algorithm, that calculates this is very important and the final result depends on its quality.
If you are interested in checking this, I can send you a simple paper, where I have explained this.

 
 Reply 
 
Alex
,
Israel, Joined Sep 2017, 8

Alex

,
Israel,
Joined Sep 2017
8
08:31 Dec-15-2019
Re: Influence of pulse frequency on measurement of material thickness change
In Reply to Wiesław Bicz at 18:08 Dec-12-2019 .

Thank you Wiesław, I will be happy to get the paper.

 
 Reply 
 
Alex
,
Israel, Joined Sep 2017, 8

Alex

,
Israel,
Joined Sep 2017
8
08:58 Dec-15-2019
Re: Influence of pulse frequency on measurement of material thickness change
In Reply to Wiesław Bicz at 18:08 Dec-12-2019 .

Wiesław,
Even if I fulfill the minimum Nyquist criterion and have 10 bit vertical range, the sampling points won't fall each on the maximum of two received pulses. Why do you state that for the time of flight measurement the sampling frequency is of much lower importance?

 
 Reply 
 
Wiesław Bicz
Engineering,
PBP Optel sp. z o.o., Poland, Joined Feb 2009, 276

Wiesław Bicz

Engineering,
PBP Optel sp. z o.o.,
Poland,
Joined Feb 2009
276
12:23 Dec-16-2019
Re: Influence of pulse frequency on measurement of material thickness change
In Reply to Alex at 08:58 Dec-15-2019 .

If you are too close to Nyquist criterion it will be necessary to take in account not only one, but many points - correlation would be the best solution. If you have sampling rate that is more than 3 times higher in many cases only two points (with good SNR): one before zero transition and one above.

 
 Reply 
 

Product Spotlight

On Demand Webinar: Using Eddy Current Array for Large Areas in Place of Magnetic Particle and Liquid Penetrant Testing

New on-demand webinar on the use of eddy current array (ECA) technology in lieu of liquid penetrant
...
testing (PT) and magnetic particle testing (MT) for large area screenings. The webinar includes an introduction to the technologies with advantages and disadvantages of each. Side by side examples are discussed including aircraft stringers, engine blades, forgings, weld plates & more. View it today.
>

Compact NDT inspection-heads for measurements with active thermography

The compact inspection head is suitable for thermographic ndt tasks. The uncooled infrared camera
...
is specially developed for NDI-tasks and offers a thermal sensitivity until now known only from thermal imagers with cooled detector. All required components and functions are integrated into the inspection-head. You will only need an ethernet cable to connect the sensor with the evaluation system.
>

TESTD-PT SYSTEM

Pulse thermography is a non-contact test method that is ideal for the characterization of thin fil
...
ms and coatings or the detection of defects. With a remarquable short test time and a high detection sensitivity, the Telops TESTD-PT is the perfect tool for non- destructive testing. With such high frame rates, it is even possible to investigate highly conductive or diffusive materials.
>

Eddy Current inspection : RotoETscan Rotating Head

The rotating head RotoETscan detects the presence of longitudinal surface or sub surface defects a
...
t high speed. The rotating head is often installed directly on the production line. It is generally dedicated to the control of long products such as tubes, bars and wires, made of ferrous or non-ferrous material. It can also be used for inspection of small parts (billets).
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window