DR vs Conventional Industrial Radiography

Dear Sir,
Please I will appreciate if anyone can give me an illustrative case on the difference between conventional radiography and Digital radiography in the following areas
Calculating space requirement and
Time to complete a job
I will appreciate also if you can explain how we can conduct DR in parallel to production weld. How is the dosage kept low and how is small area closures achieve?
Thanks in anticipation

 
 Reply 
 
Anmol Birring
Consultant,
Birring NDE Center, Inc., USA, Joined Aug 2011, 786

Anmol Birring

Consultant,
Birring NDE Center, Inc.,
USA,
Joined Aug 2011
786
01:39 Feb-19-2020
Re: DR vs Conventional Industrial Radiography
In Reply to Edafe at 15:35 Feb-18-2020 (Opening).

Best will be to contact the DR company. And are you doing DR or CR

 
 Reply 
 
Ricardo Andreucci
FATEC - Technological University of Sao Paulo, Brazil, Joined May 2018, 35

Ricardo Andreucci

FATEC - Technological University of Sao Paulo,
Brazil,
Joined May 2018
35
01:56 Feb-25-2020
Re: DR vs Conventional Industrial Radiography
In Reply to Edafe at 15:35 Feb-18-2020 (Opening).

Hi Mr. Adafe

In fact, digital radiography differs from conventional radiography only in the way of storing the obtained image. Digital radiography can be obtained using the direct DR method or indirect using the CR method. The advantage of the CR method is that it is less expensive and the screen (IP) can be curved for radiographs of curved surfaces such as tubes. Regarding radiological protection, there is no difference.
I think is a good idea to contact the supplier in order to obtain details about the software that you are using.

 
 Reply 
 
Dr. Uwe Zscherpel
Director,
BAM Berlin, Germany, Joined Jan 2010, 84

Dr. Uwe Zscherpel

Director,
BAM Berlin,
Germany,
Joined Jan 2010
84
10:14 Mar-02-2020
Re: DR vs Conventional Industrial Radiography
In Reply to Ricardo Andreucci at 01:56 Feb-25-2020 .

The difference in dose and shielding requirements between Digital Radiography and Film radiography can be huge!
Of course, if you use still the same radiation source with digital radiography than with film radiography, then there is no basic difference with digital beside a slightly shorter exposure time compared with film. BUT, if you use with digital a modern radiation source with a much smaller focal spot (and follopwingly also a lower dopse power) than the old X-ray tubes with mms of focal spot, than you will gain a lot (lower source objectdistance, lower dose poer and finally lower radiation protection requirements). Thisa was proven in the past for mobil applications when replacing Ir-192 and film by Se-75 and digital, see the literature.

 
 Reply 
 

Product Spotlight

EKOSCAN Phased Array

In order to always fit your needs, EKOSCAN can manufacture any type of UT transducer, either convent
...
ional or Phased Array. As an ISO 9001: 2015 certified company, EKOSCAN is extremely careful as far a material selection and manufacturing processes are concerned. Our probes guarantee our customers the benefits of latest innovations regarding piezo-composite, backing, impedance adaptation layer, etc. Specific probes for hostile environment: high temperature, high pressure, corrosive environment,etc. Specific probes designed to fit your specific application: optimization of every parameter to guarantee you the best detection.
>

FMC/TFM

Next generation for Phased Array UT is here now with FMC/TFM! Have higher resolution imaging, impr
...
oved signal to noise ratio, characterize, size and analyze defects better with access to several wave mode views and save raw FMC data for higher quality analysis.  Some of the benefits are:
  • Beautiful Image! Easier to understand what you're looking at
  • Completely focused in entire image or volume
  • Much easier to define setups before inspection
  • Easier to decipher geometry echoes from real defects
  • Oriented defects (e.g. cracks) are imaged better
  • See image from different wave modes from one FMC inspection
  • FMC data can be reprocessed/analyzed without going back to the field
>

Ultrasonic Probe Recharacterization Service

NDT Systems offers a comprehensive Aftercare and Recharacterization Service for all our ultrasonic
...
probes. The Recharacterization Service is fully compliant with International ASTM E1065 Standard Guide (and other applicable standards) and offers complete documentation, traceable to the ASTM E1065 Standard. For more details or to schedule Recharacterization Services contact ndtsales@ndtsystems.com
>

IRIS 9000Plus - Introducing the next generation of heat exchanger inspection.

Representing the seventh generation of the IRIS system, the IRIS 9000 Plus has nearly 200 years of c
...
ombined field inspection experience incorporated in its design. This experience combined with a strong commitment to quality and a history of innovation has made Iris Inspection Services® the undisputed leader in IRIS technology.
>

Share...
We use technical and analytics cookies to ensure that we will give you the best experience of our website - More Info
Accept
top
this is debug window